Products > Test Equipment
E36300 Series Programmable DC Power Supplies (E36311A, E36312A, E36313A)
<< < (36/54) > >>
bitseeker:
rsjsouza, are you referring to the incorrect output as compared to the display or that the supply can go beyond its maximum rating? I assumed the latter, but it sounds like HighVoltage assumed the former.

I think all my HPAK power supplies, covering several different decades, are able to go over their range. I've seen in some of the manuals where it's documented as being a feature, but accuracy is not guaranteed when going out that far.
LaurentR:
bitseeker, the E36300A cannot be programmed higher than they are specced, either through the front panel or through SCPI, so it I expect it is the former.
rsjsouza:
Bitseeker, both. I see the behaviour that LaurentR is seeing: +10% of measured current output when the set output voltage is increased above a specific value when in CC mode. That goes up to 1A set but 1.1A measured in its output.
bitseeker:
I see, so it is the aforementioned bug. Thanks for the clarification.
LaurentR:

--- Quote from: LaurentR on March 13, 2018, 05:07:10 pm ---On the series mode issue, Keysight has reproduced the issue in the US, then I got a "We are aware of this and R&D is currently working on it."

--- End quote ---

I got an update from Keysight today:
"We will have this fixed through firmware upgrade. The new firmware is expected to be available at K.com by early May."

Good news!
Navigation
Message Index
Next page
Previous page
There was an error while thanking
Thanking...