East tester ET432 vs Der EE DE-5000 Comparison
NoMoreMagicSmoke:
--- Quote from: pope on June 01, 2023, 02:34:04 pm ---Would be interesting to hear what the say.
--- End quote ---
Nothing yet. They are currently saying several months.
Def461:
Registered to comment ;D
Got my ET432 yesterday
I think I found the source of the problem
Moreover, I found that the negative reviews in the forms about 40 kHz and 100 kHz are precisely from the incorrect calibration of the device
I forcibly calibrated with a jumper from the device kit (SHORT mode), then calibrated without, but with a set of cables (OPEN mode) - the readings became completely stable in the entire frequency range within the declared parameters
UPD: 1MOhm resistor killed all my joy :(
NoMoreMagicSmoke:
--- Quote from: Def461 on June 21, 2023, 11:13:20 am ---Registered to comment ;D
Got my ET432 yesterday
I think I found the source of the problem
Moreover, I found that the negative reviews in the forms about 40 kHz and 100 kHz are precisely from the incorrect calibration of the device
I forcibly calibrated with a jumper from the device kit (SHORT mode), then calibrated without, but with a set of cables (OPEN mode) - the readings became completely stable in the entire frequency range within the declared parameters
--- End quote ---
What device were you measuring? For all of my testing I performed the open and short calibration as per the manual. Mine measured within specification up until about 10k ohms, then the resistance measurements started dropping rapidly for higher resistances.
Def461:
Updated :(
On "Big Rs" ET432 giving a lot of crap :(
NoMoreMagicSmoke:
--- Quote from: Def461 on June 21, 2023, 12:44:28 pm ---Updated :(
On "Big Rs" ET432 giving a lot of crap :(
--- End quote ---
Dang you had my hopes up that they might have fixed it...
Still nothing from the vendor. I think I give it about 10% chance they actually do anything about it.