| The Electrician:
Martin72, do you have a link to a spec sheet for the Panasonic 1000 uF cap you measured in reply #103? If not, do you know if it's a polymer capacitor, or some other type that would be exceptionally low loss? |
| MegaVolt:
--- Quote from: The Electrician on July 31, 2022, 11:10:40 pm ---Here are a couple of plots vs frequency of parameters of a 1000uF 16v Rubycon capacitor. --- End quote --- With what device these wonderful graphs were taken? |
| The Electrician:
This: https://www.hioki.com/us-en/products/lcr-meters/3-ghz/id_5772 Also Google this: IM3570E14-23M.pdf |
| mawyatt:
For those of us without a nice Hioki IM3570 we can generate some decent plots of DUT Capacitance Impedance Magnitude and Phase with a scope and signal generator. Simply sweeping the source frequency over the desired range and measuring the amplitude and phase of the voltage across across the DUT and current thru the can yield the result with a little math. Z = V/I (Magnitude and Phase), where V is voltage across the DUT and I is the current thru the DUT which can be measured by inserting a known resistor R in the signal path to the DUT. Then the DUT current is Voltage across the resistor divided by R. If you have a modern DSO like the Siglent SDS2104X+ which has a built-in Bode function you can employ this to make the nice plots directly on the screen in dB magnitude and phase of Cap DUT Z vs. frequency. If folks are interested we could start another thread on this technique, so as not to dilute this thread. Anyway, here's a plot of a 1000uF 16V electrolytic capacitor, another 10uF, 680uF and 100uF. The scale is in dB 1000 ohms and phase is in degrees. Best, |
| Martin72:
So... Today I´ve measured the both panasonic caps at work again, the menu of the ST2830 is not the clearest one. A few minutes ago I did the same again at home on the ET4410, to have it all in one post. The 100µ cap at 120Hz: ST2830 ET4410 103.16µF 103.33µF Xc -12.86 Ohm -12.84 Ohm Q 51.10 51.80 D 0.0196 0.0192 ESR 0.252 Ohm 0.250 Ohm Phase -88.88° -88.89° At 100Khz: 26.50µF 15.40µF -0.071 Ohm -0.104 Ohm 1.607 2.427 0.6185 0.4115 0.038 Ohm 0.043 Ohm -57.86° -67.63° 1000µF at 120Hz ST2830 ET4410 945.91µF 946.8µF Xc -1.402 Ohm -1.401 Ohm Q 28.60 29.04 D 0.035 0.035 ESR 0.049 Ohm 0.048 Ohm Phase -87.99° -88.03° At 100Khz 72.50µF 23.80µF -0.025 Ohm -0.067 Ohm 1.578 3.756 0.632 0.266 0.014 Ohm 0.018 Ohm -57.31° -75.20° At 120Hz they are like twins.. 8) Datasheets: 1000µF/25V FM Type: https://www.micro-semiconductor.com/datasheet/37-EEU-FM1E331LB.pdf 100µF/35V FR Type: https://industrial.panasonic.com/cdbs/www-data/pdf/RDF0000/ABA0000C1259.pdf |
| Navigation |
| Message Index |
| Next page |