The Electrician:
Martin72, have you had a chance to make measurements of the Panasonic 1000 uF electrolytic? If so, please share your results with us.
Martin72:
Hi,
Yes I had... ;)
Yesterday made measurements on the ST2830 at work and with the fixture, then the same at home with the ET4410.
At 120Hz, 1Khz, 10Khz and 100Khz.
Also the 100µF Panasonic.
Values will come, probably in the next hours..
Martin72:
Table...
Blue are the measures from the ET4410.
All measures with the fixture, when I got time I´ll make another table with the measures taken from them with kelvin clips.
Martin
The Electrician:
Can the ST2830 set the frequency to frequencies other than the standard 100, 120, 1k, 10k and 100k?
Can your ET4410 do frequencies other than the standard 100, 120, 1k, 10k and 100k?
mawyatt:
Martin has indicated that the ST2830 is a rebadged TH2830, and if so our TH2830 can be set to any frequency within 50Hz to 100KHz with sub Hz resolution at lower frequencies, and 1Hz resolution at 100KHz. So highly likely a DDS based signal source.