I re-checked with 1.10 and get the same results. Can someone confirm my measurements?
Why is the 121GW reading wrong in auto range/manual 5V range in this situation?
Can you reproduce the following measurements?
2.0V peak to peak sine wave 4kHz, No DC offset, duty cycle 50%. Both meters in AC mode. Firmware 1.07 on my 121GW with serial number EEVblog 000084
Auto mode on both meters, Fluke in highres mode:
121GW selects 5V range, reads 1.1133Vac
Fluke 28II selects 6V range, reads 0.7031Vac
manual mode on both meters:
121GW 5V range, reads 1.1133Vac
Fluke 28II 6V range, 0.7032Vac
121GW 50V range, reads 0.659Vac (and it looses the frequency measurement and displays 0.00kHz)
Fluke 28II 60V range, reads 0.663Vac
See the attached pictures.
Conducted similar measurements and can confirm similar results with regard to the frequency display.
I will post the results in the 121GW issues thread where it is more appropriate.EDIT decided to post the results here after doing some more testing and I think its not actually any bugs as such and just more like the characteristics of the meter front end.
anyway test results below........
Ok further to Candid's test I conducted a similar test of the AC range and secondary frequency display reading.
Results for: 2V p-p 4 kHz sine wave input (terminated into 50

.). My 121GW with firmware rev 1.10.
Other meters used in comparison where Fluke 87V and Fluke 8840a
Meters in Auto ranging.
121GW reads 0.7288V RMS and the secondary display show the frequency of 3.999 kHz.
Fluke 87V in hi res reads 0.7300V RMS
Fluke 8840a reads 0.7300V RMS
Meters set to manual range.
121GW on 5V
87V on 6V
8840 on 2V
All display the same as above as expected.
Meters set to next range up.
121GW on 50V, The voltage reading starts at 0.701V RMS for around 3 to 5 seconds then drops to 0.682V RMS and at which point the secondary frequency display goes to all zeros.
Fluke 87V on 60V, reads 0.690V RMS.
Fluke 8840a on 20V reads 0.7299V RMS and on 200V range reads 0.720V RMS.
additional measurement on the 121GW 500V range reads 0.47V RMS at both 4 kHz and 500 Hz (secondary frequency display is at 0), just for fun

.
NB * Further measurements of the frequency reading display show that it has a threshold of just over ~1.1 V before the secondary display starts to read frequency correctly when manual ranging to the 50V range.