Author Topic: EEVblog 121GW Multimeter Issues  (Read 886023 times)
0 Members and 48 Guests are viewing this topic.
Offline thm_w
  • Non-expert
Re: EEVblog 121GW Multimeter Issues
« Reply #1450 on: March 04, 2025, 10:52:07 pm »
If the HY3131 DMM front end IC is non-responsive, I wonder if it gives these same results? Or if it shows some kind of error code.
Clearly the STM32 is running and responding.
Online bdunham7
« Reply #1451 on: March 05, 2025, 12:58:42 am »
It works  :-//
I notice that it is warmer where you are then where he was.  Can you try putting it in the refrigerator overnight and then trying it the next day?
And if chilling it does induce the failure, could crystal X1 for U3 (the HY3131) not starting cause it to behave as shown?
Offline LampHead
« Reply #1452 on: May 20, 2025, 09:51:08 pm »
Having a similar problem with the 500mA input as well. When switching from Low_Z to Hz measurements, it gives a 'LEAD' alarm with the leads plugged in the COM and V terminals. If I plug a dummy lead into the 500mA terminal, the alarm goes away. I'm using the latest firmware 2.05, new 1.5V batteries, and I've cleaned the terminals. The meter has barely been used in the last 3 years.
Any ideas?
Offline EEVblog
« Reply #1453 on: May 20, 2025, 11:42:07 pm »
Is it just a short beep and then goes away? i.e. it's not a problem yet, but indicative of a potential problem down the road.
You might be in a high EMF environment causing a false detection issue.
Offline LampHead
« Reply #1454 on: May 21, 2025, 08:07:26 am »
Hey Dave
There's a persistent beeping with "LEAD" flashing on the screen. I'm not sure I have any major EMF emitters around, as I always use the DMM in the home office where I barely get any 4G signal :(
Do you reckon I need to replace the op-amp that senses the probes, or does it just need re-calibration?
Thanks
Offline EEVblog
« Reply #1455 on: May 21, 2025, 08:51:43 am »
In that case it's contamination of some descriptions. Could be hygroscopic moisture absorption into the connector plastics.
I did a video on this but can't find it right now.
Clean the PCB inside as a starter in case it's on the PCB.
Offline LampHead
« Reply #1456 on: May 31, 2025, 12:35:22 am »
Ok Dave, I've cleaned the PCB and placed it in a bag with some silica gel. I'll provide feedback soon. Thanks!"
The following users thanked this post: EEVblog
Offline jh15
« Reply #1457 on: May 31, 2025, 05:23:00 am »
At least you didn't put in rice. Still love my 121gw.
Tek 575 curve trcr top shape, Tek 535, Tek 465. Tek 545 Hickok clone, Tesla Model S,  Ohio Scientific c24P SBC, c-64's from club days, Giant electric bicycle, Rigol stuff, Heathkit AR-15's. Heathkit ET- 3400a trainer&interface. Starlink pizza.
Offline LampHead
« Reply #1458 on: June 19, 2025, 10:11:00 pm »
I'll be damned, it worked. I left it in the bag for weeks, pulled it out, and tried it in different modes — no beeping. Put all the covers back together and put it back in the bag again. Tried the DMM again today — still no beeping.
Winter humidity in my office was the issue, at least for me.
I think it might be useful to have an option to turn lead detection ON/OFF in a future firmware update. That way, if someone doesn't have time to dry it out right away, they can still use it.
The following users thanked this post: EEVblog, BILLPOD
