I am seeing large offsets in the data collected on my DS1054Z (SN: DS1ZA163554597).
The table below shows the offsets in the data. I used the average math function on the scope to calculate the offset.
Note that the POS dial is set to 0, except for the smallest range. Changing the POS value makes no difference in the offset and turning the AC coupling does not affect the observed offset. I checked with another scope and with a multimeter, the function generator does not account for the offset.
The attached graph shows the result when the same input is presented to both CH1 and CH2. As you can see, both channels have the offset, but the offset on Channel 1 is most severe.
Is there a procedure to calibrate the scope inputs or does the scope require repair? What can you suggest to correct this issue.

Offset (mV)
Range (V/div) Channel 1 Channel 2 Channel 3 Channel 4
10 875 630 1001 780
5 980 379 654 540
2 940 379 399 160
1 1010 391 367 184
0.5 1040 370 390 160
0.2 17 82 13 19
0.1 19.3 16 8 .6 11
0.05 17.7 9 6 5
0.02 20 7.7 7.3 4
0.01 20.4 8 7.8 4
0.005 20.8 7.7 7.7 3.3
0.002 20.8 7.6 7.7 3.1
0.001 20.5 7.6 7.8 3.1
Small shift in offset with BW limit on or off - These are for BW limit off
1X probe, normal mode
Thanks.
David