Author Topic: Enabling options on Agilent ESA series E4402B E4404B E4405B E4407B  (Read 12089 times)
Offline xymox
Re: Enabling options on Agilent ESA series E4402B E4404B E4405B E4407B
« Reply #275 on: September 24, 2020, 04:18:19 am »
Yea I have been thru the "calibration" guide.. Hehehe..  Im still not sure how to set Course and Fine in the timebase in the service menu..
I do have something off on both my units. If I give them my rubidihum clock std 10Mhz it displays slightly off center if I set the SA for 1hz res bw and 100 hz span. Also the freq counter is off a bit..
Amplitute is reading slightly off at various frequency points.
These are not much off, but with this many years on these devices, its normal to have drift.
Also if we want to swap boards around between units and maintain these over the next 10 years we just gotta be able to calibrate them.
These calibration adjustments must also be kept in some flash. If the flash goes or gets wiped, thats it. No hope of recovery.
We gotta be able to do TME.. I can't find old versions of TME. Or maybe some old different program used to calibrate.
I have sent a nice email to Ron Nersesian. He has a awesome career that goes way back with HP. Back to the good old days of HP. I have suggested that hobbyists fuel young engineers and so thier future customers could well be buying this older gear from ebay. i suggested its helpful for EoL devices that are locked away with keys to maybe get a free online key gen process. This would fuel many hobbyists into engineers and maybe future customers. It wont hurt Keysight sales as hobbyists cannot afford to buy new gear and companies using this kind of test equipment don't buy old gear like this.
Im sure I wont get a response. BUT the email did not reflect back. So I did find the right email to use for him.
WTH... Why not try...
BTW the self calibration license has provisions for a single serial number instrument. That would have to be way cheaper then a single seat.
Maybe I can get a trial license. If so, I can capture all the GPIB..AND calibrate my unit at least once..
Offline xymox
« Reply #276 on: September 24, 2020, 05:21:45 am »
OooOOo... TMA has a help file for the ESA series.. It has a lot of interesting things in it..
It *SEEMS* like you could set the freq limit and pick the model when you replace the processor board.. I believe there are hardware differences tho between models, so, its not JUST a setting. BUT its interesting tho.. Plus there are a lot of things in here that are interesting.. The part about processor initialization is interesting. http://www.xymox1.com/Misc/Utilities.pdf
All the adjustments you can do.. http://www.xymox1.com/Misc/Adjustments.pdf
I am still unsure if you can manually enter any values. It may be the TSE software forces hooking up automated bench gear and standards and then operates all that via GPIB and does the calibration fully automated.. This would not be useful. It would be better, for hobbyist use, that we could enter values manually.
Offline xymox
« Reply #277 on: September 25, 2020, 12:07:59 am »
THIS is the software I want... I think... Its old too so maybe hackable.. This can do spectrograms with full width spans..
BUT its OLD... Runs on Windows 2000.. But thats fine I have a laptop of that vintage AND there is always VM..
https://www.keysight.com/en/pd-1000004487%3Aepsg%3Apro/esa-and-psa-option-230-benchlink-web-remote-control-software?pm=PL&nid=-32406.536880458&cc=US&lc=eng
Offline smgvbest
    • Kilbourne Astronomics
« Reply #278 on: September 25, 2020, 01:20:51 am »
Far as I understand this one just need Option 230 enabled on the SA.   you can do that.
Sandra
(Yes, I am a Woman :p )
Offline smgvbest
« Reply #279 on: September 25, 2020, 01:24:04 am »
TME is Keysights premier package used to calibrate instruments and sold to calibration shops all over the world
I can see in any way Keysight is going to give that up to hobbyist.  Hobbyist are not any major part of their sales revenue.
I would love it but I'll put my money on no reply or a negative reply
Sandra
Offline apples
« Reply #280 on: September 25, 2020, 10:21:58 pm »
Just want to thank the people on this thread for their wonderful effort.
I have 3 x ESA 4402Bs that now have RF Pre-Amp which is what I needed.   One of them though is showing two distinct spikes (actually holes I suppose) in the response at 997.5MHz and 1.32GHz and when you turn the RF Pre Amp, the others don't do this.   Outside these gaps, the pre-amp is working fine.
Anyone know what could cause this?  Loose connectors ??
« Last Edit: September 25, 2020, 10:23:36 pm by apples »
Offline smgvbest
« Reply #281 on: September 25, 2020, 10:30:44 pm »
What’s the serial number on the unit?
The preamp hardware is only installed after a certain serial number
Maybe you don’t have the actual hardware ???
Just guessing but might be worth a check
Sandra
Offline analogRF
« Reply #282 on: September 26, 2020, 12:59:33 am »
first of all to figure out if your preamps are working, a simple test is to set your Ref Lvl to something small like -40dBm (no signal needs to be connected) and then turn on/off the preamp. your noise floor must jump up and down by about 15-20db depending on the gain of the preamp
i am not sure if those spikes are created by the preamp. maybe as Sandra said you dont actually have the hardware for it and it is creating garbage
otherwise the preamp is self oscillating!  :o
Offline analogRF
« Reply #283 on: September 26, 2020, 01:03:12 am »
this thread has been going off rail for quite some time. it must have been pretty much closed after the final solution. things that came after that, though interesting and useful, should have been in threads of their own.
Offline apples
« Reply #284 on: September 26, 2020, 06:40:27 am »
@analogRF -fair point.
I considered it was somewhat relevant as others are sure to use the method for this and what I'm seeing might just be a limitation of just turning the 1DS option on without having the correct re-calibration software.
For completeness, I can confirm that in my case at least the preamp IS installed and IS working as expected apart from these 'spikes'.  I'll carry on the conversation as necessary on a new thread.
Offline tv84
  • Country: pt
« Reply #285 on: September 26, 2020, 07:57:44 am »
I think we are now seeing the consequences of enabling options on Agilent ESA series so not much of a derailing here, IMHO. Of course, some themes may deserve a thread on their own.
Offline analogRF
« Reply #286 on: September 30, 2020, 06:58:11 pm »
let me pose a fresh and more useful and interesting challenge with regard to these analyzers:
how can we convert E4404B (6.7GHz) to E4405B (13.2GHz) if at all?
All boards and modules including the critical ones (attenuator and RYTHM) are common between these two
so the frequency must be limited by software only
of course calibration after such upgrade is a must and can pose a problem since the cal procedure is not possible by hobbyist
but still it would be awesome  ;D
Offline smgvbest
« Reply #287 on: September 30, 2020, 09:37:43 pm »
Service menu allow change of model and frequency range.
SERVICE/-2010/SERVICE
Other way is to program EEPROM on back of Processor card
Last if you have TME and ESA Module can do it there
Sandra
Offline analogRF
« Reply #288 on: October 01, 2020, 01:54:46 am »
What's the catch then? Could it be just the calibration thing? if it can be converted so easily then why did Agilent even sell E4404B? or why did anybody buy E4405B? Their original prices were hugely different...
Offline smgvbest
« Reply #289 on: October 01, 2020, 02:28:01 am »
IDK,   I've used it to change mine to the EMC model (same firmware),   the service menu with the right password looks to even permit you to change the serial number.  the -2010 does not permit then though.  the Serial Number and Save Serial Number are greyed out so there's additional service passwords.
Sandra
Offline mankan
« Reply #290 on: October 02, 2020, 11:19:58 am »
This can also be used in the old E44xxB ESG Signal Generators.
Just checked.
Really with the same vendor name and seeds? I cannot get it to work with a lmcrypt.exe that works for ESA. I've also tried HOST_ID variations like the VSA but with ESG and ESG-D instead. None of them produce a working code (i.e. a code for an option I already have installed).
Offline analogRF
« Reply #291 on: October 02, 2020, 11:22:39 am »
maybe it is only possible to downgrade or convert to equivalent EMC model from ESA and vice versa.
someone with E4404B should try this
Online PA0PBZ
« Reply #292 on: October 02, 2020, 11:44:14 am »
Just checked.
The ESG uses the following format:
FEATURE ABC TMOMID01 1.0 permanent uncounted 0123456789AB VENDOR_STRING=0 HOSTID=12345678
The following users thanked this post: mankan