Hello All,
I've just bought a BK Precision 8550 electronic load and I have a question regarding its current ripple. I haven't really used previously electronic loads (only made some simple resistive loads if needed for a short testing). I bought the load mainly for testing the different dcdc power supplies and LED drivers I build for my projects.
My problem is with the output current ripple: when I set the load in CC mode to let's say 50mA I can observe a ca. 10mA ripple in the load current. Let me show you an example. The test setup is the following: the load is connected to a Keysight E36313A power supply, and there is a Keysight 34465A DMM measuring the current.
Test: 5V output, 50mA CC set on the load (range 3A).
The average current measured by the DMM (10PLC): 50.195 mA. This is within the manufacturer specified ±(0.05% + 0.05% FS) CC accuracy. Below are some measurements with DMM digitizing:
Is it normal for a DC load to have this much (~10mA PP) current ripple? I plan to use the load for characterizing dcdc converters, and sometimes I want to use it in the 10-100mA range where this ripple can cause considerable error.
I also made this measurement with higher load setting like 500 mA, and even lower setting 1-2 mA (which is close to the minimum settable current). The ripple remained the same. If I set the load below 10mA this meant a ripple between ca. 0-10 mA. I tried also measuring a battery (just to be sure it's not my power supply) it didn't change anything.
Your help would be much appreciated. I can still return the load if this behavior seems to be not normal.