John_Sinha:
I recently purchased this old spectrum analyser off eBay for a few quid. What is causing this erroneous trace. I have nothing connected to the input of this spectrum analyser. Any idea why the response is not flat at the low end (under 900Hz)?

jonpaul:
normal, no problem!

Baseline displays noise floor, see specifications.

Large peak at left is mixed 0 Hz, every Analyzer has similar


read up on sweep mixer lo type RF Analyzer theory, many HP applications notes

get manual

j

807:
Yep. It's called the "zero response". Just a 0Hz marker aligned with the left hand side of screen. It's mentioned on page 3.2.1 & a few other times in the manual.

https://pearl-hifi.com/06_Lit_Archive/15_Mfrs_Publications/20_HP_Agilent/01_HP_3585A-B_40MHZ_SpecAn/HP_3585A_Operating.pdf