Author Topic: Fluke 189 fails capacitance auto range, other  (Read 1615 times)
Offline FinderbinderTopic starter
  • Country: lt
« on: March 18, 2023, 09:12:56 pm »
Actually here are few problems.
1. When I turn wheel to capacity, it shows 66pF (without leads). It is offset, but stable. Is it possible to correct offset?
2. When I measure say 470uF (on auto range), it shows ok. Then I disconnect this cap and... device starts to jump in different ranges, shows different values, include negative; and this may last infinitely or sometimes it returns to 66pF. If I connect say 40nF while it is jumping it shows 40nF correctly and upon disconnecting it returns to 66pF (always). Or I can turn wheel forth and back to capacity to return it to 66pF.
3. DC mV sometimes shows tens of uV with probes shorted, sometimes few uV, and sometimes 0 or negative values (unstable).
4. AC mV shows ~150 uV with probes shorted (stable?). Reduces few times if I keep my hand nearby multimeter (even not touching it).
5. AC V shows ~3mV with probes shorted (stable?).
Device is very clean inside, also outside.
Any suggestions?  :-//
Offline Fgrir
Re: Fluke 189 fails capacitance auto range, other
« Reply #1 on: March 20, 2023, 05:13:04 pm »
Sounds to me like everything you describe is normal.
1) I see 64pf without leads on my 189, Fluke specs 2% + 5pF accuracy in the lowest cap range, but with the following note:
using Relative mode (REL ∆) to zero residual on 1.000 nF and 10.00 nF ranges
2) I see similar behavior from cap autoranging on mine, seems to be noise pickup preventing it from ranging down.
3) Specified accuracy in the lowest mV ranges is 20uV, so might not be out of spec. Are you sure your shorting connection is stable?
4) and 5) All AC ranges have the following note:
A residual reading of 8 to 180 digits with leads shorted, will not affect stated accuracy above 5 % of range
The following users thanked this post: mwb1100
Offline mwb1100
« Reply #2 on: March 20, 2023, 06:27:10 pm »
And just to help confirm what @Fgrir said, my 189 behaves pretty much the same except:
  - for #2 when I have leads in the meter it settles to about 87nF instead of 66nF
  - for #4 my 189 shows about 70uV
Offline bdunham7
« Reply #3 on: March 20, 2023, 06:33:29 pm »
Read the specifications section of the manual?
Just now I pulled my 189 off the shelf, turned it on to capacitance and it read "0.066 nF".   :)
Online coromonadalix
« Reply #4 on: March 20, 2023, 10:44:05 pm »
same for capacitor 0.065 nF   push REL  all goes to zero
0.0028 V    ac  shorted probes
0.050 mV  ac  shorted probes
  • Country: lt
« Reply #5 on: March 21, 2023, 10:12:03 am »
Well... Uni-T UT71E shows 0 in all tests above. That is why I suspected my "new" 189 to be defective  :scared:
So my UT71E is highly superior to 189? I expected not worse behavior from 189 at least. But I got only super fast continuity advantage so far as I can tell  :palm:
btw if all 189 shows ~66pF with leads off, this means systematic error which should be (able) eliminated automatically.
...and yes, I repeated tests multiple times with different leads.
Offline Swainster
  • Country: sg
« Reply #6 on: March 21, 2023, 10:40:24 am »
Haha, here is a thread complaining about almost the exact opposite - i.e. that the UT71E is reading zero when another (as it happens another Fluke) is correctly reading non-zero: https://www.eevblog.com/forum/testgear/ut71e-current-measurement-error
The reason is that in certain ranges (commonly AC ranges), some multimeters' software just reports zero instead of a potentially inaccurate low value. Typically this threshold is 5 to 10% of the full scale reading. Other multimeter manufacturers don't bother with this 'feature' - maybe because they know that they are going to get complaints either way they handle it.
« Reply #7 on: March 21, 2023, 10:40:36 am »
i would say differently,  some meter(s) have an auto zeroing, its a way to force an zero
some members like the fact that we see really whats happening instead of using some trickery ..
that was debated here on some thread
uni-t meter doesn't impress me  loll  until you have their high end models
Offline bdunham7
« Reply #9 on: March 21, 2023, 01:29:38 pm »
Perhaps!  Find a 20-50pF capacitor and measure it with both instruments and see how they do.
The following users thanked this post: Fgrir, shakalnokturn
  • Country: lt
« Reply #10 on: March 21, 2023, 04:42:22 pm »
You are right, at that range UT71E is near useless  :-- while 189 counts almost each one pF.
Offline BeBuLamar
« Reply #11 on: March 21, 2023, 11:27:51 pm »
My Fluke 189
In capacitance probe not connected it shows 0.070 nF.
With probe shorted it shows
0.001 mV DC
0.030 mV AC
0.0035 V AC.
So it's about the same as your meter.
Offline Fungus
« Reply #12 on: March 22, 2023, 12:40:12 am »
My Fluke 189
0.001 mV DC
0.030 mV AC
0.0035 V AC.
With my 187 I get:
Capacitance mode, no probe: 0.068nF
0.001 mV DC  (ie. 1 count)
0.015 mV AC   ...but if I wave my hand near the meter it jumps up into the hundreds of mV.
0.0029 V AC.
So about the same as yours.
The plural of anecdote is "data", right?  :popcorn:
Offline Topdedcenter
« Reply #13 on: August 20, 2023, 09:59:05 pm »
Came here looking for answers regarding my FLUKE 189 that does the same thing in capacitance mode. I actually opened up my meter first, found the leaking supercap, then started browsing around the forums and found that it can simply be removed since replacements are near impossible to find at this point. But the cap has been removed and the meter still turns on and measures the same standing capacitance with/without leads attached. Thanks for all the info.
Btw, i found this meter at a yard sell for like $30 from a dude who appears to have purchased a government bid wholesale shipment. There were a couple more at the time, and i regret not grabbing them, oh well. It's a solid meter and has a much faster reading response time than the basic flukes used at work.