Recommendations for high frequency power analyzer
« on: January 06, 2017, 01:20:40 am »
Hello,

I am looking to measure the output power of electronic ballasts operating from 50 khz to 150 khz, with up to 2000 V starting voltage. My google-foo can only find analyzers rated up to 1000 p-p V.  Anyone have any suggestions for a higher rated unit?

Alternately can any of these newer oscilloscopes do the math for power calcs when voltage and current are out of sync and different wave forms?

Thanks!
 

Offline turbotemp

  • Newbie
  • Posts: 7
  • Country: us
Re: Recommendations for high frequency power analyzer
« Reply #1 on: January 06, 2017, 08:45:19 pm »
Which type of lamps are you testing and what level of accuracy do you need? it gets exponentially more difficult with increasing accuracy requirements.
 

Offline offshorebearTopic starter

  • Newbie
  • Posts: 2
  • Country: us
Re: Recommendations for high frequency power analyzer
« Reply #2 on: January 06, 2017, 11:57:12 pm »
Is 1% feasible?

Medium pressure UV. Would be nice to measure 20kW but I'd settle for 6kW.

Thanks!
 

Offline 1audio

  • Frequent Contributor
  • **
  • Posts: 307
  • Country: us
Re: Recommendations for high frequency power analyzer
« Reply #3 on: January 07, 2017, 12:52:39 am »
I assume you looked at this: http://clarke-hess.com/products/2335a/  Contact them about scaling the voltage. Possibly a simple 10X divider would get you where you need to be. You may be able to find an older version on ebay for lots less.
 

Offline turbotemp

  • Newbie
  • Posts: 7
  • Country: us
Re: Recommendations for high frequency power analyzer
« Reply #4 on: January 11, 2017, 11:04:41 pm »
Is 1% feasible?

Medium pressure UV. Would be nice to measure 20kW but I'd settle for 6kW.

Thanks!


I was checking out the specs on Yokogawa's high end meter 3000 or something and its rated 0.014 x KHz % of reading+0.3% of range
this means you're going to be at +/- 1.7% at 100KHz. Is it liquid or air cooled?

With 20kW output, you have got to have good 1-2kW of ballast loss which would allow reasonably accurate thermal measurement not too far off from above.

You'd measure the input power at line frequency which can be done to fractions of percent with a relatively cheaply leased power analyzer and use the coolant flow rate or draw entire cooling air flow through mass air flow sensor and deduct this from the input power. You could alternatively put a pass-through water heating element downstream of ballast and feed power to the element through a variac and through a second channel of the power meter.  You adjust the power so delta T in coolant across ballast and delta T across heater matches. 

Deduct this value from total input and you have the lamp wattage. 
 