SteveyG:

--- Quote from: Marco Camara on December 21, 2021, 06:46:58 pm ---My first problem is that I have a Mac. So, I needed to install a Windows 10 Virtual Machine, and installed the software which I have downloaded from Hantek web page.

The first software you have to install and configure is "Keysight Connection Expert". It's included on the software pack from Hantek. Using the USB connection, it is easy to use. You can see the instrument immediately after running the software. After that, you can send commands SCPI commands, start IO monitor and, after that, you can run Hantek HDM3000 app, which is a "remote control" of the instrument, where you can press virtual buttons, and see the readings at your computer.

For LAN connection, first you have to connect your multimeter at the same network where is your computer, and configure Network Options at Utility / IO Menu. Insert a compatible static IP address, or use DHCP. Then, after you get the IP Address at instrument, you go to Keysight Connection Expert at your computer, then "Manual Configuration", and set the same IP Address. At "Set Protocol" section, select "Socket". You must use the default port 5025, and press "Accept". After some seconds, you should see the Instrument as you have connected using USB cable.

I look forward you get it running, but if you need, just call me here, and I can try to help more.

--- End quote ---

What USB driver are you using on your PC? I just get an error 28 - It enumerates as VID_049F PID_605E which is a Compaq device if the VID is to be believed?
RS232 seems to work fine, and it spoofs a 34461A if you change the SCPI protocol.

(edit: actually you can use Zadig to install the WinUSB and it appears in Keysight Connection Expert)

jezgb:
Hi,

I've just received a HDM3065.  Quite pleased with it so far, except...  I've noticed a bit of a problem.

Since I temporarily have an Agilent 34461A on my bench, I decided to measure the same voltage source (~29V) at the same time and see how the two tracked.  Pretty good until ~36 minutes into the test the HDM suddenly showed a measurement of ~53V.  And then every 36 minutes thereafter, the same thing.  Roughly 60V, give or take, rogue reading.  Makes a bit of a mockery of the stats!  The spikes are very definitely not real, as shown by the Agilent and also tested with a 'scope.

Attached are two screenshots showing the issue.  Note the spike to the right of the Hantek graph.

Same thing on current measurement, spikes every ~36 minutes except -ve going this time showing as '-ve overload', and 2W resistance - once again higher readings than is the truth.

This smells of software to me - but I can't seem to find any updated firmware on Hantek's website.  I see there is a thread on SW for this meter but haven't had time to digest it and my skim of it filled me with dread...

The software version is 2.0.0.6(2394081348), firmware is 04.01.07.04.23-02.05-21.

Has anyone else noticed this issue?  Or do I have a unit with a pretty whacky fault?

I've emailed support@hantek.com but received no reply yet.  This, I understand, is to be expected.   :(

Such a shame because the hardware seems pretty nice.


Cheers,

         Jez.

Microcheap:
I can confirm that it's software related. I manage to reproduce and observe the same behavior.

Hantek sent me a firmware update but apparently it did not solve this issue yet. I will try to run more tests later to confirm.

ali_asadzadeh:

--- Quote ---Hantek sent me a firmware update but apparently it did not solve this issue yet. I will try do run more tests later to confirm.
--- End quote ---
Thanks for sharing, would you please your hantek contact info, because they did not answered their emails to me?

jezgb:

--- Quote from: Microcheap on April 10, 2024, 04:42:14 am ---I can confirm that it's software related. I manage to reproduce and observe the same behavior.

Hantek sent me a firmware update but apparently it did not solve this issue yet. I will try do run more tests later to confirm.

--- End quote ---

Thank you for this information.

Shame there is no fix...   Yet.  But at least I now know that sending it back for replacement is pointless.   :-+

Do you know if Hantek are aware that this issue exists?


Many thanks and best regards,

         Jez.