At first, the usual AC Volts ranges (always 50 Hz sine):
[Range]
[other meter] // [MAXCOM]
200 mV
10,0 mV // 10,0 mV
50,0 mV // 50,0 mV
100,0 mV // 100,0 mV
180,0 mV // 180,2 mV
2 V
0,100 V // 0,097 V
0,500 V // 0,497 V
1,000 V // 0,997 V
1,800 V // 1,797 V
20 V
1,00 V // 0,97 V
5,00 V // 4,98 V
10,00 V // 9,99 V
18,00 V // 18,00 V
200 V
10,0 V // 9,7 V
50,0 V // 49,9 V
100,0 V // 100,0 V
180,0 V // 180,2 V
Now the AC Amps ranges (still 50 Hz sine):
[Range]
[other meter] // [MAXCOM]
20 µA
1,00 µA // 0,95 µA
5,00 µA // 4,95 µA
10,00 µA // 9,95 µA
18,00 µA // 17,95 µA
200 µA
10,0 µA // 9,7 µA
50,0 µA // 49,7 µA
100,0 µA // 99,8 µA
180,0 µA // 179,8 µA
2 mA
0,100 mA // 0,098 mA
0,500 mA // 0,498 mA
1,000 mA // 0,998 mA
1,800 mA // 1,802 mA
20 mA
1,00 mA // 0,98 mA
5,00 mA // 4,98 mA
10,00 mA // 9,99 mA
18,00 mA // 18,00 mA
200 mA
10,0 mA // 9,7 mA
50,0 mA // 49,7 mA
100,0 mA // 99,7 mA
180,0 mA // 179,9 mA
20 A
0,10 A // 0,02 A
0,20 A // 0,06 A
0,50 A // 0,33 A
1,00 A // 0,83 A
2,00 A // 1,83 A
5,00 A // 4,80 A
This time i did some "sanity checks" against other, older meters without true-rms (which sh.ouldn't matter, right?) and with some Fluke handheld DMMs. All of them agree with each other with the exception of the Maxcom meter. As seen above, Volts is alright compared with other older meters (with lower counts). Everything BELOW the "20 A" range seems fine too.
As for the amps measurement, i used two meters in series ... mayhaps thats not "ideal"? Even though if i use any other meter with another one in series for amps, both of them always agree with each other. My source for everything is a restored Fluke 760A ... i know it's old as hell.
/edit: It does not matter if i have the MX-620 in series with another meter for comparison or standalone. If i set my Fluke 760A to 1A Output AC 50 Hz and the "pointer" reaches 10 ("search") on the analog scale i get around 1,000X AAC 50 Hz with my Fluke Meter. So far, everything is alright. If i couple the MX-620 alone to the 760A and reach my dialed-in output, it is still wrong and shows the above mentioned value for 1A (0,83 or 0,84).
Best wishes!