Products > Test Equipment

REVIEW - Rigol DS2072 - First Impressions of the DS2000 series from Rigol

<< < (359/566) > >>

EV:
Thanks Wim, it sounds reasonable!

EV:
In Pic-4 the average rise time is 1.187 ns and the average over shoot is 10.77 %. There I have used Rigols 1 MOhm 10x probe with 50 Ohm feed through terminator connected between the probe tip and generators output connector.


--- Quote from: EV on December 02, 2013, 09:10:48 am ---.... The over shoot with Rigot is also quite big. It should be under 3 %. Maybe the adaption with 50 Ohm feed through terminator is not very good.

--- End quote ---

JDubU:
Probe calibration?

Carrington:

--- Quote from: JDubU on December 02, 2013, 06:33:38 pm ---Probe calibration?

--- End quote ---
His probe have HF compensation?

Carrington:
In mine, just when the relay switches. The difference is subtle, but the two internal compensation circuits are not exactly adjusted:





Navigation

[0] Message Index

[#] Next page

[*] Previous page

There was an error while thanking
Thanking...
Go to full version