Teneyes:
@Nikwing

You may wish to try Roll mode at slow scan rate (10s/div)
then stop the scanning
and change Time Base faster to see more detail (100ms/div),
and scan though the stop display data for more detail
See : pics

Teneyes:
@Nikwing
You may wish to try triggering on each Pulse and recording the frames
Here is an Animating '.Gif' of 60 pulses at 120BPM
The playback is at 5 frames per sec so that is =  2.5 times the actual pulses rate

Teneyes:
@Nikwing
At this setting you can record 8128 Pulses

NikWing:
wow Teneyes :)

well, at first I try to measure just 5 mV to see if the gain stages work (using a 1:1 probe)
the curve is so "fat" that it's hard to see it move when I apply 5 mV to the circuit
I'm not yet to the point where I see a nicely looking heart beat like you show in your examples :o

I tried both, "default" and "last" in the settings, it happened with both (more than 1 try for both settings)
I'll give it another try :)

marmad:

--- Quote from: NikWing on June 21, 2014, 03:13:20 pm ---I tried both, "default" and "last" in the settings, it happened with both (more than 1 try for both settings)
I'll give it another try :)

--- End quote ---

Are you using FW v.3? If so, the new firmware has the new option 'Open' in the 'Power On' settings - although I have no idea exactly what it means. But you might try upgrading anyway, and see if it solves your problem.