DD4DA:

--- Quote from: lr1234 on June 25, 2015, 08:28:53 pm ---Default settings,test signal on ch1
Memory auto
Only 1Ch enabled (for 2Gs/s)
 If HighRes or Dot Mode changing time base lover as 500us cause freeze.
2Gs/s memory arithmetic’s  with bugs.
Temporary help-  enabling second channel(2Gs/s/2)

--- End quote ---

Default settings,
deactivate CH2,
Any Signal on CH1,
Aquire->Mode "Hires"
Memory Auto (is a Default setting)

Move over SweepTime 500µs to 1000µs and back - Scope freezes or hangs in a Loop.

In my Scope DS2202, the entire scope freezes only in Hires-Mode, not in normal mode as i found out for now.

Gerd
DD4DA:
Ich habe den Fehler an Rigol Deutschland gemeldet und die wiederum haben diesen an das R & D weitergegeben, nach dem Sie ihn nachvollziehen konnten. Ich muss schon sagen - das geht echt unkompliziert gut. Wieder ein Punkt der mir an Rigol gefällt. Bedenkt man die Komplexität dieses Messgerätes, so kann diese Fehlerart schon mal vorkommen. Wichtig ist es eben, dass sich der Hersteller sich diesem Fehler zur Brust nimmt. Ich kenne das auch anders.   

Ich hatte zu beginn diesen Fehler auch ohne HighRes erzeugen können. Das wiederholen des Flash-Vorgangs der gleichen Firmware erledigte den Fehler allerdings - jedenfalls habe ich ihn nicht mehr nachstellen können. Mit HighRes lässt er sich jedoch sehr gut und zuverlässig reproduzieren.
Was sich bei dieser Prozedur geändert haben könnte, vermag ich nicht zu beurteilen.
Mein DS2202 ist schon ein "Big Bang for a Buck" und gemessen am Preis bekommt man schon eine Menge Leistung geboten. Für ein R&S HAMEG muss man schon sehr viel mehr zahlen. Die I2C-Decoder Option habe ich dazu gekauft und wurde nicht entäuscht. Leider ist die Option CAN-Decode für mein Gerät nicht erhältlich - erst für die "A" Fassung. Daran hätte ich noch interesse gehabt. Derzeit löse ich es mit einem USB-Logikanalyser der diese Aufgabe auch gut meistert.

Einzig die Tastköpfe sind im 1:1 Betrieb mit 9Mhz Bandbreite sehr schmalbandig. Es gibt ja auch native 1:1 Tastköpfe die die 200Mhz schaffen oder man verwendet eben ein direktes Koaxkabel.

NUn, schauen wir mal was der kommende Update für uns bereit hält.
Muxr:
Google translate did surprisingly well for me:

--- Quote ---I have reported the bug to Rigol Germany and in turn have these passed on to the R & D, as they were able to understand him. I have to say - that's really straightforward good. Another point of what I like about Rigol. Bearing in mind this can fault occur, the complexity of this instrument before. It is important to specify that the manufacturer this error takes to the chest. I know the other way.

I had been able to start producing this error without HighRes. The repeat of the flash process the same firmware completed the mistake, however - at least I no longer have to readjust him. With HighRes he can, however, very good and reliable reproduce.
What could have changed in this procedure, I can not judge.
My DS2202 is already a "Big Bang for a Buck" and in terms of price do you get a lot of power on offer. For an R & S HAMEG you really have to pay a lot more. The I2C Decoder option I bought it and was not disappointed. Unfortunately, the option CAN Decode is not available for my device - only for the "A" version. That's what I would have been interested. Currently, I'll solve it with a USB logic analyzer of this task also copes well.

Only the probes are in 1: 1 mode with 9MHz very narrow bandwidth. There are also native 1: 1 probes that create the 200Mhz or you just used a direct coaxial cable.

Well, let's see what the upcoming update for us keeps ready.
--- End quote ---
DD4DA:
some here response that the last firmware update 3.3.SP2 has a bug in Hires mode and a fallback to older firmware seems tobe impossible - me to.
Well, this is wrong  - there is a supported way available to install an older Version of firmware. I did that and my scope works as proper as before the 3.3.sp2 Update was installed.
The DS2202 HW Rev1 (without an A) has more options to update as just install the latest firmare on a Memory stick and start the Scope oder via System Menu option. The third way looks like an Update via boot-loader, instead installed operating system.
 
The description of this procedure is attachted this message and was released by RIGOL.

Beware - Do not remove the Memorystick until the Scope is response, that the update process is finished - especialy it looks like the scope don't do anything -  be sure, it do.

The CH1 Backlight switch blink's  about 2 Minutes after the update start and the access of the Memorystick will be do a couple of times.
Do not plug-in the Memorystick into the USB recepticle at boot time, instead follow the instruction.

I had downgrade my scope to 3.3.SP1 succesfully and reset then to the default values after downgrade process. No Hires bug is left - of course.
We still wait to the next firmware release and hope this will be in better quality.



Deutsch:

Nach dem einige Mittsteiter und "Leidensgenossen" vom HIRES-Acquire Fehler in der Firmware 3.3-SP2 berichteten und nachvollziehbar war, habe ich dieses Problem an Rigol weitergegeben.
Nun suchte ich nach einer legalen und unterstützten Möglichkeit die Vorgängerversion der Firmware 3.3-SP1 wieder einspielen zu können.  Der reguläre Weg über den vor dem Start eingestecktem Memorystick mit dem Updatefile, funktioniert offensichtlich nur bei gleicher oder neuerer Version möglich.
Der Weg über den Boot-Loader des Geräts lässt aber den Downgrade zu. Er ist leider wenig kommunikativ - also es ist Gedult angesagt.
Die Prozedur ist in dem angehangenem PDF von Rigol beschrieben. Dieser Weg hat sehr gut funktioniert und ich habe mein Gerät (DS2202) wieder mit 3.3-SP1 ausgestattet. Dieser Weg ist von Rigol unterstützt da er offiziell beschrieben ist. Bitte jedoch die nötige Sorgfalt walten lassen und GEDULDIG warten  bis der Update durch ist. Zu Begin könnte man meinen, dass nicht passiert (das emine ich mit wenig kommunikativ) aber es passiert was - da kann man sicher sein. Nach etwa 2 Minuten kann man viele kurze Zugriffe auf den Stick beobacheten und kurz danach ist das Flashen abgeschlossen.
Das erkennt man an der nun nicht mehr blinkenden CH1 Hintergrundbeleuchtung, die wärend des Updates blinkt. Den USB-Stick nicht beim Starten einstecken !!. Nach Abschluß des ersten Starts müssen die Default-Werte via Storage->Default eingesteltl werden, damit die Reaktion des Skope wieder nachvollziehbar sind.

So warten wir nun geduldig auf den nächsten Update, der hoffendlich etwas besser ausfällt. 

vy 73 de Gerhard, DD4DA
DD4DA:

--- Quote from: Teneyes on July 01, 2015, 07:28:28 pm ---
--- Quote from: DD4DA on July 01, 2015, 06:44:39 pm ---ILeider ist die Option CAN-Decode für mein Gerät nicht erhältlich - erst für die "A" Fassung.

the option CAN Decode is not available for my device - only for the "A" version.

--- End quote ---
I have DS2000 Non -A and here is CAN

Load the option Key for CAN

Check what version of FW,

Ich habe DS2000 Non -A und hier ist CAN Laden Sie die Wahltaste für CAN Überprüfen Sie, welche Version von FW

--- End quote ---

Well, does the 300Mhz option works? I would assume that this needs more different parts and the HW version 2 got them. The firmware maybe works, but none of the articles in this thread confirms the reliablity of the messurement at 300mhz bandwith with hw version 1.
Your Scope shows 100Mhz, 200Mhz, 300Mhz - i owen an DS2202 that works with 200Mhz bw and this shows just 200Mhz. 

DD4DA