Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #100 on: February 09, 2023, 11:35:30 am »
Except when the manual is wrong. You have to go into the Segmented menu to find the option to change it.

Hi Dave!  When I start from a preset (little green button), it works as described in the manual.  I tried this with both the older 1.1.2 FW and the current 1.2 FW.  I can make a short video if you like, but if you hit Preset, then Acquisition -> N-single/Avg count, the value should be editable (default value is 1), and Run Single should work as expected (make a single acquisition unless you tell it otherwise).

Enabling "Acquisition -> Segmented -> Acquire Maximum" does gray out the acquisition count (since, well, maximum is maximum :)).  My guess is that this got set somehow during a previous session.  A tool tip or some other indication that this is enabled might be helpful here.

As someone who spent 20 years demonstrating loaner instruments to customers, I'm a big fan of the "Preset" button.  Can't count the number of hours I spent scratching my head trying to figure out an issue that was due to some setting that I either hadn't made or had forgotten that I had made :)

Thanks for the video!

Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8

Free online test and measurement fundamentals courses from Rohde & Schwarz:  https://tinyurl.com/mv7a4vb6
 
The following users thanked this post: egonotto, luudee

Offline pdenisowskiTopic starter

  • Frequent Contributor
  • **
  • Posts: 630
  • Country: us
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #101 on: February 09, 2023, 11:48:42 am »
Except when the manual is wrong. You have to go into the Segmented menu to find the option to change it.

Hi Dave!  When I start from a preset (little green button), it works as described in the manual.  I tried this with both the older 1.1.2 FW and the current 1.2 FW.  I can make a short video if you like

Short video :)  [Screen capture only, no audio]

Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8

Free online test and measurement fundamentals courses from Rohde & Schwarz:  https://tinyurl.com/mv7a4vb6
 
The following users thanked this post: egonotto, luudee

Online EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #102 on: February 09, 2023, 12:57:04 pm »
Except when the manual is wrong. You have to go into the Segmented menu to find the option to change it.

Hi Dave!  When I start from a preset (little green button), it works as described in the manual.  I tried this with both the older 1.1.2 FW and the current 1.2 FW.  I can make a short video if you like, but if you hit Preset, then Acquisition -> N-single/Avg count, the value should be editable (default value is 1), and Run Single should work as expected (make a single acquisition unless you tell it otherwise).

Enabling "Acquisition -> Segmented -> Acquire Maximum" does gray out the acquisition count (since, well, maximum is maximum :)).  My guess is that this got set somehow during a previous session.  A tool tip or some other indication that this is enabled might be helpful here.

As someone who spent 20 years demonstrating loaner instruments to customers, I'm a big fan of the "Preset" button.  Can't count the number of hours I spent scratching my head trying to figure out an issue that was due to some setting that I either hadn't made or had forgotten that I had made :)

Thanks, but you shouldn't have to hit preset to do that!
 
The following users thanked this post: egonotto, goaty

Online EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #103 on: February 09, 2023, 01:00:30 pm »
If only there were things called manuals where you could go and type the word "single" and have the answer in a few seconds:
Except when the manual is wrong. You have to go into the Segmented menu to find the option to change it.
They must have moved it around.
Still, the clue is there...

Sure, but I found it anyway, it just wasn't at all obvious on first use. And it's still not clear how/why it's enabled/disabled with the various related options I showed. I think the implementation is a bit of a mess for what is otherwise a great feature.
 
The following users thanked this post: egonotto

Offline pdenisowskiTopic starter

  • Frequent Contributor
  • **
  • Posts: 630
  • Country: us
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #104 on: February 09, 2023, 01:51:17 pm »
Thanks, but you shouldn't have to hit preset to do that!

Preset just returns the instrument to the default settings.  In my (sometimes painful) experience at two major T&M instrument manufacturers, it can be easy to forget about or overlook a setting, so Preset just allows you to start from a known state with default values.

Preset is so useful that most major T&M companies have some version of it -- it actually drives me crazy when I'm trying to use an instrument (even something as simple as a function generator or programmable power supply) that doesn't have some kind of preset. :)

It would be helpful (as I think a YouTube commenter said) to have something that tells you why a parameter is grayed out.  I'll definitely feed that back to our product teams.
Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8

Free online test and measurement fundamentals courses from Rohde & Schwarz:  https://tinyurl.com/mv7a4vb6
 
The following users thanked this post: Someone, egonotto, tooki, luudee

Offline Bud

  • Super Contributor
  • ***
  • Posts: 6911
  • Country: ca
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #105 on: February 09, 2023, 02:18:34 pm »
"Preset" button can be confusing if all it does is loading default settings. It should be called "Reset" instead. Presets (at least to my view) should be customer configurable.
Facebook-free life and Rigol-free shack.
 
The following users thanked this post: egonotto, tooki, Anthocyanina

Offline pdenisowskiTopic starter

  • Frequent Contributor
  • **
  • Posts: 630
  • Country: us
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #106 on: February 09, 2023, 03:07:09 pm »
"Preset" button can be confusing if all it does is loading default settings. It should be called "Reset" instead. Presets (at least to my view) should be customer configurable.

Loading default ("factory") settings is the normal behavior of preset in almost all of the (probably over 100) instruments I've used in my career. 

That said, some instruments also have a "User Preset" that basically does a preset and then loads user settings - i.e. a "customer configurable preset" [ Specified attachment is not available ]

The MXO4 supports both types :)
Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8

Free online test and measurement fundamentals courses from Rohde & Schwarz:  https://tinyurl.com/mv7a4vb6
 
The following users thanked this post: egonotto, Kean

Offline Rich@RohdeScopesUSA

  • Frequent Contributor
  • **
  • Posts: 457
  • Country: us
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #107 on: February 09, 2023, 03:07:49 pm »
Noise values ​​for one "Industry-leading system architecture: 18-bit vertical resolution/12-bit ADC" are not so exceptional low as R&S would like them to be.

For MXO4 is stated - RMS noise floor ( HD mode ??? active for bandwidth ≤ 500 MHz)

sensitivity:

HDO4: analog bandwidth   1 GHz               LeCroy HDO  1 GHz

0.5 mV/    98 µV                                                       
1 mV/       104 µV                                                         145 μVrms
2 mV/       116 µV                                                         145 μVrms
5 mV/       152 µV                                                         150 μVrms
10 mV/      238 µV                                                        155 μVrms
20 mV/      436 µV                                                        185 μVrms
50 mV/      1.01 mV                                                      275 μVrms
100 mV/    2.47 mV                                                      500 μVrms
200 mV/    4.43 mV                                                      1.75 mVrms
500 mV/    10.13 mV                                                    2.75 mVrms
1 V/div      19.96 mV                                                    4.90 mVrms

For example, every LeCroy HD Oscilloscope has in sensitivity over 2mV  significantly better values, partly 4x better!

That is not bashing, only pure facts.
This is why we post our noise figures - so people can decide what is most important to them.  In our research talking to customers, the smaller volt/div settings were the most painful for them when noisy (e.g. see the Marco Reps video above).  If you think of it as a percent of scale, versus just an absolute number (I believe LeCroy uses eight vertical divisions versus ten for us), you can see noise is much more impactful at the smaller volt/div settings.  This is also true as you move up in bandwidth - the signals are typically smaller with less margin for error, so having better noise performance on the smaller settings becomes more important.

             R&S                                              LeCroy   
Scale   RMS Noise              % of Full Scale   RMS Noise   % of Full Scale
0.001   0.000104              1.0%                   0.000145     1.8%
0.002   0.000116              0.6%                   0.000145     0.9%
0.005   0.000152              0.3%                   0.00015     0.4%
0.01           0.000238              0.2%                   0.000155     0.2%
0.02           0.000436              0.2%                   0.000185     0.1%
0.05           0.00101              0.2%                   0.000275     0.1%
0.1           0.00247              0.2%                   0.0005     0.1%
0.2           0.00443              0.2%                   0.00175     0.1%
0.5           0.01013              0.2%                   0.00275     0.1%
1           0.01996              0.2%                   0.0049   0.1%

But again, your needs may vary.  Wish we never had to make design tradeoffs, but like everyone, we do.

-Rich
« Last Edit: February 09, 2023, 03:10:38 pm by Rich@RohdeScopesUSA »
 
The following users thanked this post: nctnico, Someone, egonotto, 2N3055, pdenisowski, luudee, Detlev

Offline bozidarms

  • Regular Contributor
  • *
  • Posts: 175
  • Country: at
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #108 on: February 09, 2023, 04:07:58 pm »
Quote
If you think of it as a percent of scale, versus just an absolute number (I believe LeCroy uses eight vertical divisions versus ten for us), you can see noise is much more impactful at the smaller volt/div settings.

Thank you for your efforts.
Well, whether it's 1% or 0.1% full scale in the end doesn't matter.
If a signal is smaller than own oscilloscope noise, is and stay it  invisible.
« Last Edit: February 09, 2023, 04:09:39 pm by bozidarms »
 

Online egonotto

  • Frequent Contributor
  • **
  • Posts: 722
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #109 on: February 09, 2023, 04:11:03 pm »
Hallo,

@Rich@RohdeScopesUSA:

However, RMS is not the right measure here. Better is peak to peak or the really visible width of the noise band.


But I would be interested to know why LeCroy HDO 1 GHz at 1 V/div is so much better. Especially compared to RTA4004, even if this is 1.25 V/div in comparison to 1 V/div.


Best regards
 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 6631
  • Country: hr
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #110 on: February 09, 2023, 04:14:01 pm »
Hallo,

@Rich@RohdeScopesUSA:

However, RMS is not the right measure here. Better is peak to peak or the really visible width of the noise band.


But I would be interested to know why LeCroy HDO 1 GHz at 1 V/div is so much better. Especially compared to RTA4004, even if this is 1.25 V/div in comparison to 1 V/div.


Best regards

Every scope frontend is combination of attenuators and amplifiers...
It all ends up how you combine them in the end...
How you combine them is also influenced with needed offset ranges...

Making a scope is the ultimate art of compromise..
 
The following users thanked this post: Rich@RohdeScopesUSA, pdenisowski

Offline Rich@RohdeScopesUSA

  • Frequent Contributor
  • **
  • Posts: 457
  • Country: us
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #111 on: February 09, 2023, 04:20:34 pm »
Hallo,

@Rich@RohdeScopesUSA:

However, RMS is not the right measure here. Better is peak to peak or the really visible width of the noise band.


But I would be interested to know why LeCroy HDO 1 GHz at 1 V/div is so much better. Especially compared to RTA4004, even if this is 1.25 V/div in comparison to 1 V/div.


Best regards
I hear you - but no one specs Vpp because it can be impacted by so many other things (e.g. update rate - you'll catch the outliers more often with a faster update rate).  With that said, see @2N3055's comment above. It's really true.

-Rich
 
The following users thanked this post: 2N3055

Offline Rich@RohdeScopesUSA

  • Frequent Contributor
  • **
  • Posts: 457
  • Country: us
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #112 on: February 09, 2023, 04:22:55 pm »
Quote
If you think of it as a percent of scale, versus just an absolute number (I believe LeCroy uses eight vertical divisions versus ten for us), you can see noise is much more impactful at the smaller volt/div settings.

Thank you for your efforts.
Well, whether it's 1% or 0.1% full scale in the end doesn't matter.
If a signal is smaller than own oscilloscope noise, is and stay it  invisible.
No worries - I'm happy to help and I really appreciate your comments too.  And what you said is the challenge - pulling a small signal out of the noise.  Since that is more likely to be really hard at small volt/div settings with tiny signals, that is why we focused there. 

-Rich
 

Offline bozidarms

  • Regular Contributor
  • *
  • Posts: 175
  • Country: at
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #113 on: February 09, 2023, 04:25:55 pm »
In fact, Lecroy noise - numbers are even better than proclamated.

Quote
For MXO4 is stated - RMS noise floor ( HD mode ??? active for bandwidth ≤ 500 MHz)

How is actually, big noise by MXO4 in normal - 12 bit mode?
« Last Edit: February 09, 2023, 04:39:24 pm by bozidarms »
 
The following users thanked this post: egonotto

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28377
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #114 on: February 09, 2023, 07:27:04 pm »
"Preset" button can be confusing if all it does is loading default settings. It should be called "Reset" instead. Presets (at least to my view) should be customer configurable.
Many new instruments are.
It's a very convenient feature to return your scope to your preferred setup with a single button press rather than need to recall a previously saved Setup however they can be very useful for more specific tasks in that you can save several.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #115 on: February 09, 2023, 10:43:47 pm »
However, RMS is not the right measure here. Better is peak to peak or the really visible width of the noise band.

Peak to peak noise can greatly depend on the sample rate and waveform update rate. i.e. all thing being equal, a faster update rate scope will appear "noisier".
This is why RMS noise is used for comparison.
 

Online EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #116 on: February 09, 2023, 10:51:49 pm »
"Preset" button can be confusing if all it does is loading default settings. It should be called "Reset" instead. Presets (at least to my view) should be customer configurable.
Loading default ("factory") settings is the normal behavior of preset in almost all of the (probably over 100) instruments I've used in my career. 

Most scopes have a dedicated Default button which is different from the user Preset function. Keysight, Tek, Lecroy, Rigol, Siglent, GW Instek all call it Default.
R&S is the standout here in calling it Preset.
Yes, the Preset button on the MXO4 is actually the Default function, I would not have guessed that without trying it. "PreSet" in english means literally as it says, "pre", i.e. you have set up the configuration before you Set the function.
« Last Edit: February 09, 2023, 11:00:44 pm by EEVblog »
 

Online EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #117 on: February 09, 2023, 10:57:38 pm »
"Preset" button can be confusing if all it does is loading default settings. It should be called "Reset" instead. Presets (at least to my view) should be customer configurable.
Many new instruments are.
It's a very convenient feature to return your scope to your preferred setup with a single button press rather than need to recall a previously saved Setup however they can be very useful for more specific tasks in that you can save several.

In the case of the MXO4 the Preset button works as the Default button on other scopes. You can set it up in the setting to a user defined function though, but by default, it's Default function as per all other scopes.
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 26906
  • Country: nl
    • NCT Developments
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #118 on: February 09, 2023, 11:24:48 pm »
Yes, the Preset button on the MXO4 is actually the Default function, I would not have guessed that without trying it. "PreSet" in english means literally as it says, "pre", i.e. you have set up the configuration before you Set the function.
Maybe this is the RF background of R&S leaking into the user interface. On RF equipment there is no default button but a Preset button to set the equipment to default settings.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: tooki

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28377
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #119 on: February 10, 2023, 12:13:31 am »
Yes, the Preset button on the MXO4 is actually the Default function, I would not have guessed that without trying it. "PreSet" in english means literally as it says, "pre", i.e. you have set up the configuration before you Set the function.
Maybe this is the RF background of R&S leaking into the user interface. On RF equipment there is no default button but a Preset button to set the equipment to default settings.
Yes I see lots of Preset on anything RF whereas scopes all seem to use Default.
Either way, factory Default/Preset is the default option and User definable Defaults/Presets in a couple of forms available too. Some assign user settings to a Default/Preset while in older gear this could also be done as a Setup, just not as quickly/easily recalled.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #120 on: February 10, 2023, 08:00:22 am »
The MXO4 locked up on me again while shooting a clip on the digital logic analyser.
Similar lockup result to last time.
The bug last time was that I operated the buad rate decode control too quickly. We'll see what it was this time.

 
The following users thanked this post: egonotto

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16649
  • Country: 00
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #121 on: February 10, 2023, 08:23:18 am »
Peak to peak noise can greatly depend on the sample rate and waveform update rate. i.e. all thing being equal, a faster update rate scope will appear "noisier".
This is why RMS noise is used for comparison.

Shouldn't it be standard deviation to eliminate any DC offset?
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16649
  • Country: 00
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #122 on: February 10, 2023, 08:27:29 am »
The MXO4 locked up on me again while shooting a clip on the digital logic analyser.
Similar lockup result to last time.
The bug last time was that I operated the buad rate decode control too quickly. We'll see what it was this time.



Weird that all the menus and stuff keep working. They need to add a watchdog to the acquisition engine/FPGA.
 

Offline goaty

  • Regular Contributor
  • *
  • Posts: 204
  • Country: de
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #123 on: February 10, 2023, 09:16:15 am »
Software is not easy to design. I think a lot of 'traditional' companies did and do struggle with good software design.
I remember the remote control of a Canton speaker for TV which was ridculous as well as the 'Rademacher' remote control system for blinds.
In the end, R&S is also probably has not the 'Google'-level of software development...
 
The following users thanked this post: tooki

Online EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: First picture on EEVblog of the new R&S MXO4 series oscilloscope :)
« Reply #124 on: February 10, 2023, 09:40:23 am »
Peak to peak noise can greatly depend on the sample rate and waveform update rate. i.e. all thing being equal, a faster update rate scope will appear "noisier".
This is why RMS noise is used for comparison.

Shouldn't it be standard deviation to eliminate any DC offset?

Yes, also called AC RMS.

 