| hafrse:
Hello, I am trying to adjust the flatness on my 8564E but did not work: If I choose a band #3 and try to adjust the DAC at the specific frequency 12.4Ghz , previous DAC value was 3501 , if I change to 3256 and can see the trace within the flatness menus at the wanted level , save and exit. Now measuring the same frequency @12.4Ghz still shows level value which was on DAC 3501! going back to the flatness cal menu and set back band #3 12.4Ghz , the level is correct within that menu and the DAC value is 3256 (the new value) . I tried values like 100 and 4995 , DAC value will be saved but trace still the same at 12.4Ghz. any clue what that can be? However, in the low band mixer (9kHz-3GHz) it is possible to adjust and see the results without any problems. Where do I find the flatness compensation DAC ? Thanks for any advice! George |