Products > Test Equipment
Fluke 101 calibration
Chris Learing:
--- Quote from: mqsaharan on August 30, 2024, 11:59:51 am ---I hope it'll help you realign your meter.
--- End quote ---
Fantastic. Worked like a charm. Thank you very much. Though I can't begin to imagine how you worked that out by trial and error! Thanks again.
mig_skyfighter:
I hope it'll help you too) :)
mqsaharan:
Thank you mig_skyfigher for the calibration manual.
I couldn't find it on fluke's website. Would you like to share where did you find it?
Also, I guess I was wrong with 4.500V input. Or perhaps, 101 needs 4.500 while 107 needs 5.000V input. I don't own a 107 so I cannot check.
Fungus:
--- Quote from: mqsaharan on August 30, 2024, 05:33:11 pm ---Thank you mig_skyfigher for the calibration manual.
I couldn't find it on fluke's website. Would you like to share where did you find it?
Also, I guess I was wrong with 4.500V input. Or perhaps, 101 needs 4.500 while 107 needs 5.000V input. I don't own a 107 so I cannot check.
--- End quote ---
I won't be difficult to try them both and find out.
mig_skyfighter:
My source of information is this video.