Isn't the minimum resolution limited due to the True RMS converters ?
From Fluke 179 spec sheet with regard to the 600mV range:
"Unspecified input levels on the lowest ranges are:•AC voltage: below 5% of 600 mV AC, or 30 mV AC"
From Fluke 80's series V manual:
"AC conversions for Model 87 are ac coupled, true rms responding, and valid from 3 % to 100 % of range, except 400 mA range (5 %
to 100 % of range) and 10 A range (15 % to 100 % or range)"
i would imagine that the Brymen has similar restrictions. 3% of 5V is 150mV (+40 digits and .8% accuracy if I read the specs correctly).
rsjsouza, I did notice that your readings were quite similar between meters. Perhaps because you were using a sine wave?