Has anyone measured Flukes 15B or 17B's burden voltage on any current range? There's no mention about on the manual provided on Fluke's website.
I am especially interested in the 4A and 10A range.
-
From the manual, get the shunt resistance, and use ohms law?
-
My 17B shows 0.05 Ohm in A range, 2.23 Ohm in mA range, 101.3 Ohm in uA range for series resistance at the sockets. Each of its supplied leads shows 0.05 Ohm. Measured with MS8218 delta zeroed. From there maybe you can estimate.
-
From the manual, get the shunt resistance, and use ohms law?
I don't think that's specified in the manual.
-
My 17B shows 0.05 Ohm in A range, 2.23 Ohm in mA range, 101.3 Ohm in uA range for series resistance at the sockets. Each of its supplied leads shows 0.05 Ohm. Measured with MS8218 delta zeroed. From there maybe you can estimate.
Thanks Salas!
So roughly it would be 0.2V on the 4A range (0.05V per Amp ) and 0.4V on the 10A range (0.04V per Amp)? Of course excluding the leads.
-
Have a look at this:
Fluke 17b multimeter and DC Current Measurement (https://www.youtube.com/watch?v=ei34lIp5qSs#)
-
Have a look at this:
Fluke 17b multimeter and DC Current Measurement (https://www.youtube.com/watch?v=ei34lIp5qSs#)
Interesting, actually it was me who contacted that user and asked him about current measuring capabilities, but I did not check his channel after that.
So if I do another rough estimate with 0.040 Ohm on the 4A range it would be 0.16V (0.04V per Amp). This is similar to Fluke 289 (burden voltage according to Fluke's manual).