Re: East Tester ET4410 Desktop LCR Meter
« Reply #125 on: December 25, 2023, 11:36:24 am »
Finally, I tested whether it was due to my device - it's not, Testcontroller finds it without any problems.


Offline Andrey_Ak

  • Contributor
  • Posts: 42
  • Country: kz
Re: East Tester ET4410 Desktop LCR Meter
« Reply #126 on: December 25, 2023, 12:28:41 pm »
It works for me:

 
The following users thanked this post: Martin72

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: East Tester ET4410 Desktop LCR Meter
« Reply #127 on: December 25, 2023, 01:04:42 pm »
Hm,

You also have a com port displayed, but the program can't find anything for me...
Which OS do you have? I am running Win10.

Offline Andrey_Ak

  • Contributor
  • Posts: 42
  • Country: kz
Re: East Tester ET4410 Desktop LCR Meter
« Reply #128 on: December 25, 2023, 01:10:09 pm »
When starting the program, I can select any COM port.
Windows 7 x64  v.6.1.7601
 
The following users thanked this post: Martin72

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: East Tester ET4410 Desktop LCR Meter
« Reply #129 on: December 25, 2023, 01:43:15 pm »
Quote
Windows 7 x64  v.6.1.7601

Aha... :D
@All:
If I use the compatibility mode for win7, then it works!
The colors in the illustration are somewhat unfortunate, perhaps this can be changed.
But it works perfectly and the ET4410 is recognized and displayed without any problems.

Online JohanH

  • Frequent Contributor
  • **
  • Posts: 626
  • Country: fi
Re: East Tester ET4410 Desktop LCR Meter
« Reply #130 on: December 25, 2023, 01:44:17 pm »
Got the program working in wine after "winetricks allfonts". Set a COM1 port in the registry.

Still can't access the device. I can select COM1 in the program.

I can't find any specs if it requires a specific port speed. Tried also swapping RX and TX on the cable but no go.
 

Offline w.v.s.

  • Regular Contributor
  • *
  • Posts: 87
  • Country: de
Re: East Tester ET4410 Desktop LCR Meter
« Reply #131 on: December 25, 2023, 02:30:47 pm »
This is how my program looks like (at least mine shows more contast  :-DD ):
Left plot is time series for primary measured value, next to it a histogram of that value.
First line: Time, last measured primary value, average and standard deviation.
Second line: Selection of primary unit, measured secondary value, real part of impedance, imaginary part of impedance.
Third line: Selection of secondary unit, frequency.

Not the best C++ coding style. Currently running with 22.04.1-Ubuntu. Also talks with GPIB, LXI, and some USB devices.
 
The following users thanked this post: thm_w

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5897
  • Country: ca
Re: East Tester ET4410 Desktop LCR Meter
« Reply #132 on: December 25, 2023, 03:24:44 pm »
 :-+ 

but man    i hate all the black  thingy / fashion / trend,   way too hard to see,    use light gray  instead,  if your room get too dark or darker in a sense, you loose the contrast effect and almost see nothing ??

black is not for me  loll  even on instruments front panels ...

kudos
 

Offline Andrey_Ak

  • Contributor
  • Posts: 42
  • Country: kz
Re: East Tester ET4410 Desktop LCR Meter
« Reply #133 on: December 25, 2023, 03:31:11 pm »
Aha... :D
@All:
If I use the compatibility mode for win7, then it works!
The colors in the illustration are somewhat unfortunate, perhaps this can be changed.
But it works perfectly and the ET4410 is recognized and displayed without any problems.



Why does your interface look prettier than mine?
I have a question mark after the value "ROUT".
All the selection fields are curved, the colors are also different.
And I can't change the value of the "ROUT" value, there are 100 and 30 ohms, I can choose but they don't change on the device, it always shows 100 ohms.



 

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: East Tester ET4410 Desktop LCR Meter
« Reply #134 on: December 25, 2023, 04:13:18 pm »
Hi,

Quote
Why does your interface look prettier than mine?

No idea, I´ve changed nothing.

Quote
I have a question mark after the value "ROUT".

And me a "0" instead of  \$\Omega\$, makes ist not better.

Quote
And I can't change the value of the "ROUT" value, there are 100 and 30 ohms, I can choose but they don't change on the device, it always shows 100 ohms.

Now you´ve mentioned, same here... :(
I have examined the exe file with Uni-Extract, the LCR programme itself is written in Delphi, perhaps something can be changed/corrected there.

Offline pope

  • Frequent Contributor
  • **
  • Posts: 289
  • Country: pl
Re: East Tester ET4410 Desktop LCR Meter
« Reply #135 on: December 25, 2023, 06:22:41 pm »
This software looks pretty hopeless to me :)
 

Offline MathWizard

  • Super Contributor
  • ***
  • Posts: 1431
  • Country: ca
Re: East Tester ET4410 Desktop LCR Meter
« Reply #136 on: December 25, 2023, 08:43:26 pm »
So as they are built right now, what's capping the max frequency at 100kHz ? Say you wanted to extend it to 1 or 10MHz, would much have to be added or changed ? Someone would have to hack the MCU or whatever's in itm to add the extra range.


I'm interested in trying that with my LCR meter, but I'm over my head with the MCU.
 

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 5833
  • Country: de
  • Testfield Technician
Re: East Tester ET4410 Desktop LCR Meter
« Reply #137 on: December 25, 2023, 09:05:08 pm »
Quote
So as they are built right now, what's capping the max frequency at 100kHz ? Say you wanted to extend it to 1 or 10MHz, would much have to be added or changed ?

It won't work that easily.
I had drawn a part of my ET4410 as a circuit diagram and if I remember correctly, the circuit corresponds exactly to the one on the left side(lower corner) of this picture:

For frequencies above 100kHz it becomes much more complex, simply increasing the frequency is not enough, the whole circuit has to be adapted/replaced.
It's not worth it.
On the contrary, I started with the circuit diagram to find out how to optimize the existing circuit so that it works up to 100kHz, because I have my doubts that it does.
Edit: Pic taken from the Impedance Measurement Handbook (Agilent):
https://wiki.epfl.ch/carplat/documents/rcl_agilent.pdf
« Last Edit: December 25, 2023, 09:21:30 pm by Martin72 »
 
The following users thanked this post: thm_w, MathWizard, w.v.s.

Offline w.v.s.

  • Regular Contributor
  • *
  • Posts: 87
  • Country: de
Re: East Tester ET4410 Desktop LCR Meter
« Reply #138 on: January 04, 2024, 10:42:22 pm »
Did a small face lift on my ET4410. Might appear boring, but I did not like the look of that orange  ;D
Proper red would also be acceptable but was not on stock.
 
The following users thanked this post: Martin72

Offline thm_w

  • Super Contributor
  • ***
  • Posts: 6378
  • Country: ca
  • Non-expert
Re: East Tester ET4410 Desktop LCR Meter
« Reply #139 on: January 04, 2024, 11:32:48 pm »
Received the ET4410
Internals are basically the same as the ones Martin posted other than the micro. Instead of STM32F103 we have a more powerful N32G455. Though the meter is probably running the same code for the most part. The UI front panel uses N32G457.

Parts seen:
- N32G455 144MHz 32-bit 5MSPS ADC, 1MSPS DAC
- SGM3003 analog switch
- MS8052M high speed 5V 250MHz op-amp

The factory CAL routine mentioned above (hold SET when powering on) requires: 1.4k, 4.64k, 15k, 47.5k, 174, 25.5, 6.81, 2.05, and "standard impedance" whatever that is.

To disassemble:
- Remove two large philips head screws at the back of the case
- Remove the tilting bale by rotating it, then pop out
- Sheet metal case will slide off

edit: jlink can connect to the N32G455RE, but of course the flash is locked so no easy way to read it out.
« Last Edit: January 05, 2024, 12:06:38 am by thm_w »
Profile -> Modify profile -> Look and Layout ->  Don't show users' signatures
 

Offline w.v.s.

  • Regular Contributor
  • *
  • Posts: 87
  • Country: de
Re: East Tester ET4410 Desktop LCR Meter
« Reply #140 on: January 04, 2024, 11:38:33 pm »
Received the ET4410
Internals are basically the same as the ones Martin posted other than the micro. Instead of STM32F103 we have a more powerful N32G455. Though the meter is probably running the same code for the most part. The UI front panel uses N32G457.

If your device is still open: Did you look if there is any activity on the through-holes for the front-usb, where no connector is mounted?
« Last Edit: January 04, 2024, 11:40:52 pm by w.v.s. »
 

Offline thm_w

  • Super Contributor
  • ***
  • Posts: 6378
  • Country: ca
  • Non-expert
Re: East Tester ET4410 Desktop LCR Meter
« Reply #141 on: January 05, 2024, 12:24:06 am »
If your device is still open: Did you look if there is any activity on the through-holes for the front-usb, where no connector is mounted?

There is an unpopulated connector JP2 + U10 at the top, presumably goes to the GPIB connector on the back. JP3 is SWD port. Then JP5 which is right above the USB hole, has 5V, Gnd, then two other pins that presumably go to the 32G457 OTG port. But there is no activity or voltage on the two middle pins, so I don't think any connection is implemented in the software its running.
Profile -> Modify profile -> Look and Layout ->  Don't show users' signatures
 

Offline thm_w

  • Super Contributor
  • ***
  • Posts: 6378
  • Country: ca
  • Non-expert
Re: East Tester ET4410 Desktop LCR Meter
« Reply #142 on: January 05, 2024, 12:42:42 am »
For whatever reason HPOT output has some distortion on it whereas HCUR signal does not. I guess when they are connected its normal so it doesn't matter?
Output is:
- 2V at 10kHz and 1.8V at 100kHz with no load.
- 0.67V and 0.57V with 50R load.
« Last Edit: January 05, 2024, 12:52:49 am by thm_w »
Profile -> Modify profile -> Look and Layout ->  Don't show users' signatures
 

Offline w.v.s.

  • Regular Contributor
  • *
  • Posts: 87
  • Country: de
Re: East Tester ET4410 Desktop LCR Meter
« Reply #143 on: January 05, 2024, 09:56:21 am »
If your device is still open: Did you look if there is any activity on the through-holes for the front-usb, where no connector is mounted?

There is an unpopulated connector JP2 + U10 at the top, presumably goes to the GPIB connector on the back. JP3 is SWD port. Then JP5 which is right above the USB hole, has 5V, Gnd, then two other pins that presumably go to the 32G457 OTG port. But there is no activity or voltage on the two middle pins, so I don't think any connection is implemented in the software its running.

Thank you for checking this!
 

Offline mawyatt

  • Super Contributor
  • ***
  • Posts: 3263
  • Country: us
Re: East Tester ET4410 Desktop LCR Meter
« Reply #144 on: January 05, 2024, 03:04:05 pm »
For whatever reason HPOT output has some distortion on it whereas HCUR signal does not. I guess when they are connected its normal so it doesn't matter?
Output is:
- 2V at 10kHz and 1.8V at 100kHz with no load.
- 0.67V and 0.57V with 50R load.

HPOT and LPOT are LCR Meter Signal High Z Sensing Inputs, HCUR is the Drive Signal Output Source and LCUR is the Drive Signal Sink, usually at virtual ground as forced by a Transimpedance amplifier.

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 