Author Topic: Fluke 189II/AN (289) Service Manual  (Read 5609 times)
Offline Black PhoenixTopic starter
  • Country: hk
« on: March 09, 2023, 03:25:43 am »
Well, I was repairing a computer from a friend who works in a repair facility here in Shenzhen and come around to a folder with the name Fluke.
Inside I found the 87V/AN service manual and...
The 189II/AN, mostly known as 289 nowardays. I never saw this manual online in Google searches or any website. So for the good of the community here it is. And yes, it includes the block diagram, schematics plus calibration procedure.
https://www.eevblog.com/forum/testgear/fluke-189iian-(289)-service-manual/?action=dlattach;attach=1734632 - part 1
https://www.eevblog.com/forum/testgear/fluke-189iian-(289)-service-manual/?action=dlattach;attach=1734635 - part 2
Instructions:
Download both, remove the .txt extension (so rename them so they end in .001 or .002) and extract with 7-zip.
« Last Edit: March 10, 2023, 01:44:52 am by Black Phoenix »
Online coromonadalix
Re: Fluke 189II/AN (289) Service Manual
« Reply #1 on: March 09, 2023, 04:03:34 am »
Do note U9 has been rendered  phantom in the maual   loll
Thks a lot
  • Country: hk
« Reply #2 on: March 09, 2023, 04:16:42 am »
Thks a lot
It is not U9, the measurement processor, which is a Fluke branded item, the reason why?
Offline zrq
« Reply #3 on: March 09, 2023, 02:04:19 pm »
Note that by hacking the PDF a bit you can see through the mist above the U9.
The following users thanked this post: jemotrain, Black Phoenix
  • Country: hk
« Reply #4 on: March 09, 2023, 02:41:33 pm »
What, it means it is something added via the PDF password protection?
I don't know because I done everything via mobile. Just copied the file to my phone via BT, upload it to the cloud and posted the link here. I still didn't got it in my Laptop to check it, been a busy day, less than a hour away from home now in transportation.
« Last Edit: March 09, 2023, 05:38:40 pm by Black Phoenix »
« Reply #5 on: March 09, 2023, 03:56:47 pm »
in what way ?? would be nice to share in PM ?
  • Country: hk
« Reply #6 on: March 09, 2023, 05:13:19 pm »
Well now that I arrived and after the hint from ZRQ, there is definitely something under it...
Offline zrq
« Reply #7 on: March 09, 2023, 05:35:14 pm »
Just try any PDF editor, should be a simple white box placed above the text layer. I'm still on my work computer without one, so just spotted the dodgy redaction and posted.
  • Country: hk
« Reply #8 on: March 09, 2023, 05:37:32 pm »
Yes I realized that there is a lot of redactions, on the top centre white space of the circuit there is a ton more...
Need to check later, too late here now.
It doesn't surprise me to be sincere. Some years ago, way back in the old Flukecommunity.com, I found a unreleased update for the FVF software by playing with the link of the old one.
Basically by changing the version number at the link.zip.html it started downloading a non yet released updated version of the FVF.
It was at the time reported by me in the forum, with one of the Fluke Corporation employees (all of them were mods there) confirming that such link was real, updating the main website one week later with it.
But before someone confirming it it was close to 4 weeks after my findings, with most users there using the latest version of the program while in the main Fluke website, the Support tab for the program, only had the update before it. So they took some time to confirm and update such link.
Unfortunately the web.archive.org doesn't have any reference for it, or it has the post in question, but the snapshot available is years before such happened.
So going by what I found just by playing with numbers in the name of a file, it doesn't surprise me that Fluke hidden part of the schematic in a way that someone was able to get that something was there.
With means that there is 2 versions of this file. This one who is provided to the US Government as the 87V/AN one, and the official service centre one who probably doesn't have the redactions (just my hypothetical deduction).
« Last Edit: March 10, 2023, 05:52:53 pm by Black Phoenix »
« Reply #9 on: March 09, 2023, 07:34:21 pm »
Can confirm; open it in LibreOffice Draw and you can remove the white tile redactions:
« Last Edit: March 09, 2023, 07:36:58 pm by AVGresponding »
The following users thanked this post: coromonadalix, Black Phoenix
« Reply #10 on: March 09, 2023, 10:03:00 pm »
thanks a lot,     lots of troubles to hide a few things ...   they where some hidings in other pages,  not sure i saw all of them
p45-46
strange  coverting the whole pdf document in libreoffice draw  ODG   is way smaller   loll
« Last Edit: March 09, 2023, 10:10:49 pm by coromonadalix »
« Reply #11 on: March 09, 2023, 10:05:11 pm »
p47-48
I dont know about pcb changes and FW  for the old 189 II with the "now" 289 vesrion
maybe someday  we will have some photos
« Last Edit: March 09, 2023, 10:12:34 pm by coromonadalix »
Offline alm
« Reply #12 on: March 09, 2023, 10:33:52 pm »
Nice discovery about the redactions! I opened it in a trial version of Acrobat Pro and removed the white rectangles on pages 35 and 37. Please let me know if I missed any. I had to split it in two posts to stay under the attachment limit. Download both, remove the .txt extension (so rename them so they end in .001 or .002) and extract with 7-zip.
The following users thanked this post: lowimpedance, msquared
Offline alm
« Reply #13 on: March 09, 2023, 10:34:46 pm »
Part 2
Offline mwb1100
« Reply #14 on: March 09, 2023, 10:52:38 pm »
Is that available for download too?
Offline alm
« Reply #15 on: March 09, 2023, 10:58:05 pm »
I imagine the 87V service manual is the same available here:
https://xdevs.com/doc/Fluke/87V/
The following users thanked this post: mwb1100, dophuc
  • Country: hk
« Reply #16 on: March 10, 2023, 01:05:19 am »
Yup it is the same old one that's been around for some time. I checked it and that's why I didn't copy it.
So now what we need is to inject the pages without reactions back on the original PDF and we have a more complete version. Should I delete the original link I posted in the first post?
Later I'm going to download the Lible Office and do the changes myself and put a new link on the top post. Thank you for all of you for finding something I didn't realized before.
« Last Edit: March 10, 2023, 01:13:54 am by Black Phoenix »
Offline alm
« Reply #17 on: March 10, 2023, 01:23:01 am »
I posted the complete version.
  • Country: hk
« Reply #18 on: March 10, 2023, 01:38:45 am »
Didn't saw, sorry. Being on mobile causes this kind of problems sometimes.
Top post edit with your links and instructions. Now I just need to upload the new version to xdevs.com.
Again thank you for everyone who found that and helped. Now waiting for C&D from the Fluke Corporation or the US Government...  :P
Also looks like I need to spend a little more time in that computer searching. Who knows what more can I find in terms of Service Manuals, right?
They had some for most Apple Products, and some Huawei too, but the Fluke Folder was the one who caught my attention.
« Last Edit: March 10, 2023, 01:57:32 am by Black Phoenix »
« Reply #19 on: March 10, 2023, 06:47:35 am »
 :-DD
« Reply #20 on: March 10, 2023, 05:27:15 pm »
@ALM
@mwb1100
The 87v  are not the same models at all
You must talk about the 87IV / 89IV   who is the same as the 187 and 189,  BUT not the same as 187II and 189II who are the 287 and 289 predecessors
Fluke just loved to F#$^  a few things  loll
Offline alm
« Reply #21 on: March 10, 2023, 05:36:38 pm »
@ALM
@mwb1100
No, no one here has been talking about the 87IV/89IV/187/189. Black Phoenix found two service manuals, one for the 189II/289, and one for the 87V. The 87V manual has been already available from xdevs.com for a while.
A service manual with schematics for the 87IV/89IV/187/189 would also be great, but I'm not aware of this being available to the public (yet).
Offline EEVblog
« Reply #22 on: December 28, 2023, 04:53:01 am »
Unzipped direct PDF for reference.
The following users thanked this post: Gertjan, coromonadalix, TheDefpom, gamalot, shakalnokturn, NA5B, mwb1100, AVGresponding, lern01