Post by: czullo on December 13, 2023, 01:59:00 pm
No matter what mode of measurement I choose, be it voltage or current, resistance or even temperature, all measurements are heavily skewed and unstable. Has anyone had a similar problem with this meter and knows how to fix it? I am the second owner so warranty repair is not likely to be an option.
Title: Re: FLUKE 289 all measurements are wrong
Post by: armandine2 on December 13, 2023, 03:10:24 pm
deleted post - not relevant
Title: Re: FLUKE 289 all measurements are wrong
Post by: J-R on December 13, 2023, 11:26:08 pm
The two photos posted so far don't appear to show a problem in my opinion.  Mine does the same thing in VAC mode if I'm near it with no leads connected.  Do you have more examples of the issue but with test probes connected to something?
Title: Re: FLUKE 289 all measurements are wrong
Post by: BeBuLamar on December 13, 2023, 11:44:14 pm
If you don't have the probes connected or connected and not shorted together or connect to a voltage source then the display is normal. Also if you live in the USA you can still have the 289 repair under warranty even if you're the second owner. Fluke doesn't ask for proof of purchase when you send it in for service.
Title: Re: FLUKE 289 all measurements are wrong
Post by: Kim Christensen on December 13, 2023, 11:50:35 pm
No matter what mode of measurement I choose, be it voltage or current, resistance or even temperature, all measurements are heavily skewed and unstable.

Getting a reading on AC volts like that isn't too surprising.
What does it read when measuring current with no leads attached?
Title: Re: FLUKE 289 all measurements are wrong
Post by: tonycstech on September 28, 2024, 04:27:35 am
My 289 has an issue with resistance.
I am reading 1.2K circuit (semi conductors BGA chips and micro processors) with any chinese meter i get 1.2K.
Fluke gives me some 200-300 ohms.
It does read other circuits normally.

I noticed that it uses different voltages for different range of resistance measurements but they say the same thing.
Title: Re: FLUKE 289 all measurements are wrong
Post by: parawizard on September 28, 2024, 07:54:16 am
My 289 has an issue with resistance.
I am reading 1.2K circuit (semi conductors BGA chips and micro processors) with any chinese meter i get 1.2K.
Fluke gives me some 200-300 ohms.
It does read other circuits normally.

I noticed that it uses different voltages for different range of resistance measurements but they say the same thing.

Are you saying you are measuring in circuit? Probably you are turning on something in the circuit which gives you a different result than other meters?