Author Topic: Rigol DG1022Z calibration  (Read 1346 times)
Offline pauleddTopic starter
  • Riesige Gepanzerte Luftschiffe
« on: January 10, 2021, 05:46:59 pm »
Hi
I just noticed my DG1022Z is a bit off in ARB mode when I select just DC output. It reads 1.2mV when it should be 0V.
I wanted to take a look into the calibration menu but it is password protected and "12345" from other RIGOL product did not work.
Does anyone know the factory password or do I need to contact RIGOL.
human being - without Windows® - excuse my bad english
Offline manupthehills
Re: Rigol DG1022Z calibration
« Reply #1 on: January 10, 2021, 09:14:20 pm »
The password is 123456
You can calibrate all the "cal item" but don't touch the high frequency flatness 'cause there's a bug and it will screw the instrument. However, it can be recovered by loading the calibration default, but then you'll be back at square zero...
« Reply #2 on: January 11, 2021, 09:11:48 am »
Thanks, works. I guess I only need to do the OFFSET items...
I just found a calibration manual:
https://www.rigolna.com/calibration/  ;D