I went to a car boot sale today and found this little gem: A Fluke 8040A DMM.
For a few Euros it changed hands and was added to my collection.
It is a handheld/bench DMM 41/2 digits and was made in 1974.
Power is provided either via a plug pack/charger or via four built in NiCd C size cells.
When I got it to my lab I first took the NiCd pack out to see if there was any damage from
leakage. Sadly there was, although it obviously wasn't the fault of the currently installed cells.
They seem to be fine.
One of the battery contacts was corroded off. I replaced it with a contact from an old portable
tape recorder.
So next test: Let's apply power from the bench PSU. Sadly nothing happens, the display shows nothing.
I could hear some high pitched noise from inside, though. Like a switch mode supply. Ok, what's going on
here? Now I had to open it up.
Isn't that beautiful? More like a work of art then electronics, really. All components are through-hole, of
course. And all the IC's are standard CMOS 40xx types. Except one, wich is a custom Fluke type.
And sure enough, there really is a small switch mode supply inside.
Luckily I was able to find a manual online.
After a few measurements I found that all the supply voltages were present and correct. The SMPS is
an up converter generating +/- 11,5V and +/- 6,5V from 4,8V.
The manual had some suggestions what might be the problem when the display shows nothing, one
of wich was worrying: The custom LSI IC could be broken. A repair wouldn't be possible then. I don't
for one moment believe Fluke could still supply that one!
There were also several other IC's that could be responsible. One was a 4049 that is used as a 1MHz
clock generator together with a 1MHz crystal. I poked around there a bit with the scope probe and found
some very dodgy looking signals. The clock was not looking right. So I replaced the 4049 with a new one.
Power on....and hey! The display is back!
And it works brilliantly.
David