| Products > Test Equipment |
| Fluke 8050A calibration |
| << < (6/7) > >> |
| CDaniel:
Half range should be fine , and if for some reason it's not perfectly linear here is a good point for compromise |
| maginnovision:
If it weren't for COVID I'd offer to let you use my calibrator... Perhaps later. Assuming 91343 is your zip we're practically neighbors. |
| bob91343:
Yes that is my zip. Where are you? |
| maginnovision:
Near the mission police station. |
| JimS:
I was working on a 8050 about 2 weeks ago. It had been on a shelf unused for around the last 3 years if not more. The ohm's with a shorting bar was off on every other ohms range. A good spray cleaning of the switches and it was ok after that. |