CDaniel:
Half range should be fine , and if for some reason it's not perfectly linear here is a good point for compromise
maginnovision:
If it weren't for COVID I'd offer to let you use my calibrator... Perhaps later. Assuming 91343 is your zip we're practically neighbors.
bob91343:
Yes that is my zip. Where are you?
maginnovision:
Near the mission police station.
JimS:
I was working on a 8050 about 2 weeks ago.
It had been on a shelf unused for around the last 3 years if not more.
The ohm's with a shorting bar was off on every other ohms range.
A good spray cleaning of the switches and it was ok after that.