I have bought over 5 of these Fluke DMMs, both 8505A and 8506A as well as one CT model from the U.S.A.F. All of them were less than 150 USD and some were as low as 80 USD with shipping included... Out of all of them, only one was a "parts unit" with a bad Controller module.
Paying more than $250 for a non CT model is a waste (the CT units have a bunch of modifications to lower noise and, for the 8506A, extend the input range of the AC voltage measurement). The 850xA are large, heavy, can have annoying transformer noise, and are very slow with the averaging mode on. The 100mV range is also pretty bad--noisy and quite drifty as these do not have an auto-zero circuit. Honestly, there are a lot better vintage meters from Schlumberger and a few other European TE manufacturers that would likely be cheaper for you to buy in the EU.
I keep my 8506A since it is an interesting anomaly using Flukes recirculating remainder ADC and custom thermal converter IC. I also have an HP 3456A (circa 39th week 1990) that I keep because its so stable.
All that said, the 8505/6A have pretty good 10V and 1kV DC ranges. The 100V range could have been improved greatly by adding an extra tap on the custom laser trimmed divider network that was added to later units. But Fluke decided not to do that, which was pretty dumb, considering they ended up using a PLD to control the relays. The FW wouldn't have needed to be updated--the GAL26CV12B could have correctly set the gain for the newer cards and be backwards compatible with old FW. Then there are improvements to the Active Filter that could have been done to improve the LSD. But...it is what it is: a decent vintage DMM if you can get a good deal (<150 USD with GPIB/Serial and Ohms/Current Shunt modules). The AC performance isn't really that important, as it is unlikely that you can get it calibrated for a reasonable price. And I have no doubt that every one of these meters sold on the second-hand market is out of calibration on the AC ranges.