Some time my Fluke 87 III displays an Over-Load on volts scale without anything connected. A simple tap on the rotary switch is enough to fix it. Bad contacts?
(https://i.imgur.com/75qS354s.jpg) (https://i.imgur.com/75qS354.jpg) (https://i.imgur.com/8T0nYVks.jpg) (https://i.imgur.com/8T0nYVk.jpg)
Alexander.
-
Certainly sounds like it.
Pull it apart and rub them with spit, that usually does the trick.
-
See
http://mrmodemhead.com/blog/fluke-83-iii-rotary-switch/ (http://mrmodemhead.com/blog/fluke-83-iii-rotary-switch/)
"It basically works and seems to be calibrated correctly, but occasionally when set to measure Ohms, the display will flicker a bit and only display “OL”, regardless of whether the leads are connected to anything, or even shorted together. A slight tap on the rotary switch will cause it to start working correctly again. Odds are, there’s a minor problem with the internal switch contacts."
-
DeOxit D-100L. Apply directly to switch contacts. Do not use a spray cleaner.
-
I refer you to time 12m02s in this video:
https://www.youtube.com/watch?v=u2-ot2vWLxI (https://www.youtube.com/watch?v=u2-ot2vWLxI)