Author Topic: Fluke 8840a - AC voltage reading low  (Read 1264 times)

0 Members and 1 Guest are viewing this topic.

Offline StlHifiTopic starter

  • Newbie
  • Posts: 1
  • Country: us
Fluke 8840a - AC voltage reading low
« on: July 01, 2017, 05:53:47 am »
Got this meter off eBay with the cal button's sticker torn off.  :-\ I plugged the probes into the 122 VAC wall, verified by my HP 3468b that is within spec. The Fluke read 133 mV! I tried calibrating ohms and DC volts, that worked well. I get "error 41" when I try to calibrate VAC using the HP as reference. I can calibrate the ohms and VDC just fine, after erasing the calibration. AC voltage remains elusive.  :scared:

What might be wrong, and what should I do? The self test is fine on the Fluke!
 

Offline xwarp

  • Frequent Contributor
  • **
  • Posts: 367
  • Country: us
Re: Fluke 8840a - AC voltage reading low
« Reply #1 on: July 02, 2017, 03:04:23 pm »
Pull the unit out if it's cabinet and make sure that you see a single wire, (I think it's a red wire), going from the AC board to the main board. If that is not connected, you won't get a reading. If it is connected, then you might have problems with the AC board.

I think I have a spare board somewhere around that we might could work something out on.
 