Products => Test Equipment => Topic started by: SharpEars on September 20, 2015, 02:45:51 am
-
Is it possible on the Fluke 8842A multimeter to get rid of a small current offset (say about 30 microamperes or so)? This appears when the unit is fully warmed up and in current measurement mode and there is nothing attached to it. I was looking through the manual and short of going through the entire calibration procedure for all functions, I see no way to get rid of a fairly small current offset (i.e., to zero out the reading when there is no current flowing through the instrument).
If anyone has any thoughts or knows of a way to do this, please help me out.
-
Just use math REL function to null readout?
-
Does pressing the offset button work?
-
Yep, that's exactly what the OFFSET button does. See section 2-29 in the manual "Offset Measurements".
-
Does pressing the offset button work?
Yep, that's exactly what the OFFSET button does. See section 2-29 in the manual "Offset Measurements".
I must have been blind, the Offset button did the trick! I guess I was looking for a Zero button and tried to search for zero/zeroing in the manual, as it is called on my other DVMs. Apparently, the word offset didn't click in my head as "compensating for an existing offset."