Products => Test Equipment => Topic started by: stanward on March 24, 2021, 06:24:15 am
-
Hello,
I have a Fluke 88V that used to work fine until I noticed the battery leaked. Cleaned it up, worked fine after that. Then a day later, the display will only show "Hi 2" all the time. I opened up the case again and nothing looks bad in there. What could be wrong with it?
-
The 88V will display "Hi 2" if you power it up with the "Hz %" button pressed. Perhaps the button is stuck down or the contact area is contaminated?
-
Thank you for your quick response. Let me open up the meter again, as I didn't notice anything odd last night.
The 88V will display "Hi 2" if you power it up with the "Hz %" button pressed. Perhaps the button is stuck down or the contact area is contaminated?
-
Thank you J-R! You were right! Visibly there was nothing awry with the contact pads but I used a polymer eraser to clean the contacts and any of the thru-hole vias associated with the two contact pads and the meter now works again!
Thank you!
The 88V will display "Hi 2" if you power it up with the "Hz %" button pressed. Perhaps the button is stuck down or the contact area is contaminated?
-
Glad to hear it's fixed!