Products => Test Equipment => Topic started by: stanward on March 24, 2021, 06:24:15 am

Title: Fluke 88V only displays Hi 2
Post by: stanward on March 24, 2021, 06:24:15 am
Hello,

I have a Fluke 88V that used to work fine until I noticed the battery leaked.  Cleaned it up, worked fine after that.  Then a day later, the display will only show "Hi 2" all the time.  I opened up the case again and nothing looks bad in there.  What could be wrong with it?
Title: Re: Fluke 88V only displays Hi 2
Post by: J-R on March 24, 2021, 07:39:23 am
The 88V will display "Hi 2" if you power it up with the "Hz %" button pressed.  Perhaps the button is stuck down or the contact area is contaminated?
Title: Re: Fluke 88V only displays Hi 2
Post by: stanward on March 24, 2021, 06:45:24 pm
Thank you for your quick response.  Let me open up the meter again, as I didn't notice anything odd last night.

The 88V will display "Hi 2" if you power it up with the "Hz %" button pressed.  Perhaps the button is stuck down or the contact area is contaminated?
Title: Re: Fluke 88V only displays Hi 2
Post by: stanward on March 24, 2021, 07:00:07 pm
Thank you J-R!  You were right!  Visibly there was nothing awry with the contact pads but I used a polymer eraser to clean the contacts and any of the thru-hole vias associated with the two contact pads and the meter now works again!

Thank you!

The 88V will display "Hi 2" if you power it up with the "Hz %" button pressed.  Perhaps the button is stuck down or the contact area is contaminated?
Title: Re: Fluke 88V only displays Hi 2
Post by: J-R on March 24, 2021, 09:06:15 pm
Glad to hear it's fixed!