Tokar: try 0.019d?
v0019d Device after calibration, and autoset. Sensitivity 500mv/div. We increase the DC voltage until the limit is reached, press autoset. And, lo and behold, the signal (2.4v 500mv/div) without limitation, expanded to full screen (0:29). Setting time by pressing autoset is ~5s.
We increase the signal to the limit (5v). We press autoset, the range switches to 1v/div. But the limitation remains. Still the same 4.5 cells as reported earlier (#1697). We increase the voltage to 7V, press autoset. The range changes to 2.5v/div, the limit disappears (1:50).
Then I gradually increase the voltage, and at 16V I get a transition to 5v/div (3:39). But the "0" marker is in the center of the screen. I increase the voltage again, checking the status each time by pressing autoset. There is a limitation at 22V, but there is no offset of the “0” marker (5:20). At 31v we get a marker shift, and, as a result, the signal limitation is removed (7:30).
https://youtu.be/2mIM5b9XkMI