Great feedback and insights as usual, really appreciated! It’s a little hard for this old Bodger to keep up with it all, but that’s not a complaint, it just takes me some time to go through things and take it all in.
I hear everyone loud and clear on solving the 5V rail, and I concur, don’t get me wrong. I'm afraid that my ignorance of this kind of equipment and circuitry, has my pea brain struggling to comprehend how an electrical gap in the attenuator section
couldn’t be contributing to the symptoms I’m seeing. But I’ve been studying the attenuator schematic and I think I’m starting to understand why I shouldn’t be concerned. (Any insights on that front would likely be useful though). I think I’ll just continue to try and sniff out the problem with the attenuator cable out of circuit entirely moving forward, and just get over it.
That said, fabricating the missing attenuator parts is going to be a real task, in and of itself. That’s why I’m going to start the process of finding a shop that can do it. I’m thinking not just any 3-D print set up can do this properly. A precision scan will have to be done, the right plastic will need determining, etc. In any case, I won’t pull the trigger on fabrication until the 5V rail mystery is solved, but there’s some leg work to do on this aspect of the repair.
Tony_G – Thanks for the printing offer, much appreciated! We’ll see how it goes.
I think I’ve beat the components along the H string to death at this point, I’m convinced there’s no component level issue along there.
All of those IC pins are showing a positive voltage of approximately the same value, 1.6 to 1.7V. By the schematic, I think it should be 5V or thereabouts. I'm thinking the issue lies downstream of this section. I’ve been looking at service sheet 15 for the next place to investigate, giving some “side eye” to chips U1, U22, U24 and U32.
Great test suggestions swainster, thanks! As for the touch test, I’ve done a good bit of that and haven’t found anything finger blistering yet, even after prolonged power up, but it’s a thing I’ll be constantly doing. Also, I measured resistance across all of those caps statically a while back and the caps were consistent in their readings, giving me some confidence that they’re solid.
Anyway, I’m taking a brief sidetrack with a cute little Spectrum Analyzer I acquired recently for a couple of days. It needs some mechanical stabilizing before I can shelve it.
Then I’ll get back to the HP. I’ll report back with any new developments as they occur, and always love the comments folks! Thank you!
JRH