Bottom line: While it’s now clear to me what the RTB is doing, the bigger concern here is the SDS. After a trigger, it misses something like 10 possible triggers before its ‘arming’ again. For a trigger signal as slow as 1 kHz, this is quite strange. Perhaps, like Martin72 suggests, there might be a serious bug here that is still waiting to be fixed?
Hi.
Could you check something on Siglent? If you go into Acquistion menu, do you have Slow or Fast mode set there? You need to be in the dot mode also..
A small suggestion, if I may. To avoid confusion, may I suggest a wording specificum: a time that scope needs to be ready for a new trigger event can be called
retrigger time. We can also use a term
rearm time or even a
blind time to signify time that scope is "blind" to new events while trigger engine is being rearmed for another go. I say that because saying that "scope is missing triggers" can be misconstrued as scope being ready for trigger but it didn't recognize it properly.
Difference is that first one is operating specification of the scope and the other one (missed triggers) is defect, a bug.
A bit of pedantry, I know, but nevertheless not unimportant sometimes.
As Someone correctly said, phosphor emulation works by virtue of sampling short bursts of acquisitions synchronised with screen refresh rate. Keysight also does it on certain timebase settings. Even KS 3000T that is clocked at more than 1 milion triggers per second does it at some settings.
But this is nothing new, really. This is pretty much only thing KS does better in Infiniivision scope series (Megazoom IV based). Whole architecture is based around this feature. With all the compromises that stem from it, like very small memory, use of decimated buffers for all measurements, all time interpolation with no user control etc. These are pretty much speciallistic scopes made to emulate something similar to analog CRT scope triggering performance as primary design goal.
Both R&S and Siglent, with their long memory architecture are closer to "analytic" scope type, where you capture longer sequences and then analyse it. In fact, we can argue that these scopes with long memory can achieve "zero" blind time for bursts of, say, 100ms. This way of thinking, though, does need for user to actually adjust way how they are using scopes, because it is slightly different to how you would use CRT scopes many people are used to.
That is why I always keep repeating that, for instance, KS 1000X series are not very capable scopes (they have very limited capabilities compared to even scopes many times cheaper) but are good for people that don't need advanced features (user wants to to only look at waveform on the screen and maybe use cursors and basic measurements) and that want as good as possible CRT emulation.