Products > Test Equipment
“Blown” Fuses in Several New Multimeters
(1/9) > >>
Locator:
I’ve run across something that is very perplexing to me. The other day, I accidentally checked 110v AC house voltage with my Fluke 117 on the AC current setting. Of course, the outlet sparked and it slightly damaged the tip of the probe. The multimeter did not turn off, but I thought I probably blew the fuse… Only time I’ve ever done this, but it was a moment of getting distracted and I kicked myself for getting distracted….

I opened it and tested continuity on both fuses. One of them did not work with continuity and also showed “OL” when checking resistance. Both tests were done with a brand new Fluke 179. Okay, I suppose I need to order a fuse from Fluke (I’m assuming 110v AC will blow the 10a fuse?)

Then, just out of interest, I then opened up the brand new Fluke 179 and tested both fuses in the 179 (out of the meter). One of them was open - no continuity. (Checked with a new Fluke 87V.) What?? This was a brand new meter and the only thing I tested with it, for function was continuity, the resistance of the leads, a couple of AA alkaline batteries and the AC house current. IN the AC voltage setting - NOT in the current setting. I wondered if maybe it was a return to Amazon and someone else blew the fuse and sent it back, I don’t know.

Then, for fun, I checked the 87V. both fuses showed continuity.

Then, I checked two Brymen 257’s (actually, Greenlee DM-510a’s). Both fuses were okay in these.

Then, I opened a Brymen 827 (it was a Greenlee DM-830a) and the larger fuse was open - no continuity. I purchased this new from Amazon a few years ago and it has led a pampered life - barely used. And, never any current. It was only tested with continuity with the leads a few years ago and kept in the case.

Now, I had a couple of fuses on the workbench that I took out of a dead Klein MM6000 I had for years. It also had a pampered life and had only checked continuity a handful of times and the resistance of the leads. I took it out of the case a few months ago and the display was dead. I figured it was because it was a Klein, as I have had two other Kleins die a long time ago. I removed the two fuses - 1000v and 600v and kept them for spares. I checked them for continuity after doing all the other checking and BOTH of them were open. No continuity. Checked with a perfect Fluke 87V.

And last one - I have an Amprobe AM-570, that I also bought new (all of these meters were bought new). It has only checked DC voltage and continuity. Also used it to check 110v house AC to check its’ accuracy compared to the Flukes. BOTH of its’ fuses were open - no continuity.

Any thoughts on how I ended up with all of these dead fuses? They are all the factory fuses that came in the new multimeters and the only one that I evidently damaged was the Fluke 117. These things are the poster children for perfect condition multimeters.

I know it sounds like these meters have been in the hands of a careless user, but they’re hardly used and I am the only one that has touched them.

Shouldn’t ALL fuses in multimeters show continuity when tested?
Jeroen3:
Did you do this with the wires you used to measure 120V of AC current? Because I've seen flimsy wires get damaged by it.
Genuine fluke wires   probably   not, but you should check anyways.

(I've also evaporated the probes once with exactly the same mistake as you, only on a bigger socket, not fun)
wraper:

--- Quote from: Locator on September 05, 2024, 06:41:35 am ---Then, just out of interest, I then opened up the brand new Fluke 179 and tested both fuses in the 179 (out of the meter). One of them was open - no continuity. (Checked with a new Fluke 87V.) What?? This was a brand new meter and the only thing I tested with it, for function was continuity, the resistance of the leads, a couple of AA alkaline batteries and the AC house current. IN the AC voltage setting - NOT in the current setting. I wondered if maybe it was a return to Amazon and someone else blew the fuse and sent it back, I don’t know.

--- End quote ---
Do you mean you tested AC voltage? Selected switch position does not matter for this. If the probe was plugged into 10A or 400mA socket, it will blow the fuse.
Fungus:
Meters don't have to be in "amps" setting to blow a fuse.

It's enough just to have the cable in the "A" connector when you try to measure voltage.

You don't need voltage to blow a fuse, just amps. You can blow a fuse with 1V if there's enough amps.
Locator:

--- Quote from: Jeroen3 on September 05, 2024, 07:59:02 am ---Did you do this with the wires you used to measure 120V of AC current? Because I've seen flimsy wires get damaged by it.
Genuine fluke wires   probably   not, but you should check anyways.

(I've also evaporated the probes once with exactly the same mistake as you, only on a bigger socket, not fun)

--- End quote ---

If you mean, did I test each of the fuses with the damaged leads, no - they were set aside so I could clean them up, later. The tests were done by two different sets of Fluke silicone-insulated leads that I had ordered to replace the factory leads that come with the multimeters.
Navigation
Message Index
Next page
There was an error while thanking
Thanking...

Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod