Author Topic: GDM-8351 DMM calibration procedure  (Read 2041 times)
Offline kakciTopic starter
  • Country: hu
« on: April 30, 2019, 09:23:54 pm »
Hello!
I'm looking for the calibration procedure on this DMM or service manual if anyone has it.
Offline coromonadalix
Re: GDM-8351 DMM calibration procedure
« Reply #1 on: May 01, 2019, 07:53:30 am »
Question,  did you ask Gw Instek ??
  • Country: hu
« Reply #2 on: May 01, 2019, 08:12:13 am »
Yes heres the answer:
"IF you find that your unit is giving false readings and it has to be adjusted/repaired, then this can only be done by GW representative."
I was asking for the calibration procedure.
Trying to ask here since someone succeeded with the GDM-8251 calibration procedure and posted it here. But the 8351 doesnt have an USB Type-A port.
« Last Edit: May 01, 2019, 08:16:48 am by kakci »
Offline PeZe
  • Country: cz
  • HW Developer
« Reply #3 on: May 19, 2024, 09:42:52 pm »
Hello everyone,
Has anyone made any progress with the calibration procedure for this multimeter?
I just got it second hand and would like to do a calibration/adjustment.
So far what I have been able to find is that the procedure from the older model which had a dedicated USB port for the calibration key cannot be used as it is missing on this model.
I have also asked GW instek support for the procedure but the reply was clear - they will not share it unless you are an accredited laboratory or authorised distributor. :-//
Online Martin72
  • Testfield Technician
« Reply #4 on: May 19, 2024, 10:37:04 pm »
Is that necessary, how did you find out?
Do you have the appropriate means for such a calibration?
If not, I would have the meter calibrated externally if I were you.
« Reply #5 on: May 20, 2024, 03:45:40 am »
must be deon by scpi commands or dedicated software ...
maybe decompile a fw upgrade file to see the inner commands ????  if such a file is available ?
in last resort   send it to a cal facility ...
Offline PeZe
  • Country: cz
  • HW Developer
« Reply #6 on: May 20, 2024, 08:21:06 pm »
Yes, DC values are off by 2 digits on some ranges.
I have at my disposal calibrated higher count DMMs and accurate voltage/current calibrator - we are used to calibrating uncritical equipment at work by ourselves.
So it is not a matter of equipment or metrology skills.
Of course sending it to a certified lab for adjutment is an option, but for my home development I do not exactly need acredited protocol for it... Especially it is a pity to spend the money when I have the equipment available to do it myself...
I wonder why Instek is going this way as other manufacturers are going the open way and some of them provide calibration details directly in the manual...
I was just curious if anyone has been successful in obtaining the procedure since the last discussion in 2019, but I get it, these DMMs and not as popular as KEITHLEY or KEYSIGHT etc...
Online Kleinstein
« Reply #7 on: May 21, 2024, 08:09:29 am »
The specs allow for quite some tolerance, including at least 5 counts. So 2 counts off from the right value would still be well in spec.
For home use there could still be option to note down the difference / correction factor and for the few measurements that one really minds apply this.
If one has a better meter, chances are one would use that meter for the critical measurements anyway.