Hello, I have a Giga-tronics 8541C and a 80320 sensor. Everything was fine until I said to reset the unit to "default settings" so I went to Service Menu and hit the CLEAR ALL MEMORY option. After this I got the "A UNCALIBRATED" on the display so I connected the sensor to the 50MHz calibrator connector and hit the CAL/ZERO button but after around 3 seconds it returns to the same "A UNCALIBRATED" text on the display. I looked in the Service manual and it says the the CLEAR ALL MEMORY option will erase also the the amplitude calibration data performed by the Power Sweep Calibrator. I opened the power meter and I checked the voltages on the A1 board and +5V is spot on at +5.00V, -15V is at -15.02V and the +15V is at +14.97V I measured the 50MHz out and it is around 49.7MHz but the service manual says it should be between 49 and 51MHz. Also it says the power output is only good for 15 seconds and after it starts to drift. How to restore the power meter to a working state? Do you have any tips? Am I doing something wrong? Many thanks! |
| squadchannel:
It is possible that you have deleted important calibration data. If so, it would need to be recalibrated by a repair service. It will also cost money. |O However, it should not normally be possible to delete it intentionally. |
| rplabs:
--- Quote from: squadchannel on August 11, 2024, 02:13:25 pm ---It is possible that you have deleted important calibration data. If so, it would need to be recalibrated by a repair service. It will also cost money. |O However, it should not normally be possible to delete it intentionally. --- End quote --- I was thinking the same after I read the service manual but that would be very strange to allow this feature. I saw that in the same service menu I can set the calibrator coefficient % but it needs to be calculated after measurement with an HP 432A and a 478A thermistor mount. I ran more tests from the "Service Menu", maybe some can check their 8451C if it has similar values? Sensor 80320A is connected via cable to the power meter and also to the calibrator connector. Sensor "A" DVM -> 1114 mV Range 3 Under "Hardware Tests" there is a test called "5 VOLTS" and it shows 38xx (xx is changing between 17 and 21). I measured all 5V testpoints on both A1 and A2 boards and they are OK. Does this test means I have somewhere 3.800V ? |
| Fungus:
--- Quote from: rplabs on August 11, 2024, 02:32:47 pm ---Under "Hardware Tests" there is a test called "5 VOLTS" and it shows 38xx (xx is changing between 17 and 21). I measured all 5V testpoints on both A1 and A2 boards and they are OK. Does this test means I have somewhere 3.800V ? --- End quote --- No. All readings are done by comparing the incoming voltage with a precision internal reference. You then multiply the ratio of those two by a scale factor to get the number displayed on screen. That "scale factor" is the calibration info that's been lost. |
| rplabs:
--- Quote from: Fungus on August 11, 2024, 07:11:56 pm ---No. All readings are done by comparing the incoming voltage with a precision internal reference. You then multiply the ratio of those two by a scale factor to get the number displayed on screen. That "scale factor" is the calibration info that's been lost. --- End quote --- Do you know how to add the "scale factor" for the 8541C ? Or how is it calibrated in the cal lab? |