I'd used one of my own 50 Ohm feedthrough terminators and got these results.
A few minutes ago, I did repeat the test using one of yours, and I've got 1.5ns risetime using ETS and the cursors.
But using the signal generator, the BW is less with your terminator, and the amplitude was wrong (had to set the generator to a lower level to get the same 200mV rms). So, I measured your terminator, and it's 100 Ohm

This works with the pulse gen, since it's directly connected to the terminator and the scope (no cable at all), but not with the sig gen, the cable impedance mismatch creates some bumps in the frequency response.
Tested your 100 Ohm terminator at my fastest scope (1GHz TDS580) with the pulse generator: 16% overshoot, 0.53ns rise time.
My 50 Ohm terminator: 28% overshoot, 0.33ns rise time
Another 50 Ohm terminator: 31% overshoot, 0.4ns
Internal 50 Ohm termination: 11% overshot, 0.29ns (posted this somewhere in the Leo Bodnar pulse gen thread)