Author Topic: GSP-930 GW Instek Spectrum Analyzer not starting up  (Read 397 times)
Offline SixtenTopic starter
« on: February 09, 2025, 03:33:36 pm »
With risk of creating a duplicate , I don't know if my post ended up in the right category  ???
Anyway - I have this GSP-930 GW Instek Spectrum Analyzer that won't start.
Pressing the on/off-switch starts up all voltage rails and they all are ok.
On the main board containing a FPGA I connected a UART to see if the boot-sequence could get me some leads and according to attached dump it seems that the FPGA doesn't manage to start correctly.
A far as I understand the FPGA (XC3SD1800A-4CS484LI) is loaded by an externally memory which means if it's faulty or have a bad soldering underneath it should be possible to replace without having it programmed.
I know this model have kind of a "well known" issue, since I've seen several 930 with this same boot-problem and the only solution is to ship it for repair.
Anyone that's familiar with this issue or hints what to check?
Online nctnico
Re: GSP-930 GW Instek Spectrum Analyzer not starting up
« Reply #1 on: February 09, 2025, 03:57:34 pm »
It looks like the data in the NAND flash is corrupted. Try to contact GW Instek and if there is some way to recover the unit using a USB stick with a fresh firmware image on it so the NAND gets re-flashed. I don't think there is anything wrong with the hardware so don't 'repair' what isn't broken.
« Reply #2 on: February 10, 2025, 06:28:46 am »
Thanks a lot for your suggestion nctnico.
It sounds reasonable and is for sure worth a try.
I do however don't had that success to get a reply from GW Instek support last time, but I'll do another try.
Offline duster
  • Country: cn
« Reply #3 on: February 17, 2025, 03:05:30 pm »
Do you have any methods? I have got two gsp-930 with the same nand flash booting problem, can you start them? Thanks!!!
Offline duster
  • Country: cn
« Reply #4 on: February 18, 2025, 02:40:05 pm »
i think here is the answer form our bilibili: copy the bin from a good gsp 930, then replace the bad one! it worked in this up video: 【固纬GSP-930频谱仪维修记录-哔哩哔哩】 https://b23.tv/tNsesos
good luck!