David Hess:
Source impedance is a major problem with wide bandwidth differential probes and a major limitation of performance.  If you are probing a differential transmission line, then it may be better to build a test point into the circuit with 50 ohm outputs to connect directly to the oscilloscope which can then do the differential measurement itself.

joeqsmith:

--- Quote from: David Hess on August 17, 2016, 09:53:14 pm ---Source impedance is a major problem with wide bandwidth differential probes and a major limitation of performance.  If you are probing a differential transmission line, then it may be better to build a test point into the circuit with 50 ohm outputs to connect directly to the oscilloscope which can then do the differential measurement itself.


--- End quote ---

 :-// I guess that depends on what you consider a major problem.   My old DC coupled diff probes are 300 ohms at 4GHz.  Tip capacitance is a 1/10th of a pf. 

David Hess:

--- Quote from: joeqsmith on August 17, 2016, 11:37:02 pm --- :-// I guess that depends on what you consider a major problem.   My old DC coupled diff probes are 300 ohms at 4GHz.  Tip capacitance is a 1/10th of a pf.

--- End quote ---

How much source impedance mismatch are you dealing with though?  A 300 ohm differential probe is not intended for general purpose measurements; it requires closely matched source impedances to maintain common mode rejection ratio.

This Tektronix article, Making Single-ended Measurements with a Differential Probe, refers to the problem on page 11:

A 50 ohm source impedance mismatch for a single-ended signal measurement has only a minor effect on the DC CMRR of a probe with a 50 Kohm probe input impedance.  As the signal frequency increases however, the probe input impedance begins to decrease and eventually this 50 ohm source impedance mismatch for a single-ended measurement can become quite significant. High frequency AC common mode voltage transients that may be largely rejected by the relatively high CMRR in a differential measurement may show a noticeable effect due to degraded CMRR in a single-ended measurement.

tronde:

--- Quote from: JohnG on August 13, 2016, 03:37:19 pm ---I'm wondering if anyone has a working design or knows of one for a DIY wideband (DC to > 1 GHz) DIY differential probe. I've search for this on and off, and found some attempts, but have not found a successful design. I have come across references to a 2 GHz design published in Elektor July/Aug 2015, but I don't have the article so I don't know if this is a working design or not. I'd buy the issue if I had some indication that this was indeed a working design.



--- End quote ---

I don't want to break Elektor's copyright, but this is the specs:

Technical specification
• Attenuation: 10:1 with a differential signal and 50 ? termination in the ‘scope
• Differential input resistance: 5 k?
• Single-ended input resistance: 2.5 k?
• Output resistance: 50 ?
• Bandwidth: 1.9 GHz (–3 dB)
• Rise/fall time: 300 ps
• Power supply: ±8 to 12 V DC.

They also write:

For readers who are interested the author
offers ready-to-use and tested PCB modules, also a kit consisting of case, RF cable
with BNC connector and power supply
lead with plug. Further information from:
alfred_rosenkraenzer@gmx.de.

joeqsmith:

--- Quote from: David Hess on August 18, 2016, 06:31:04 am ---How much source impedance mismatch are you dealing with though?  A 300 ohm differential probe is not intended for general purpose measurements; it requires closely matched source impedances to maintain common mode rejection ratio.

--- End quote ---

Differential digital, LVDS etc.