Author Topic: Hacking the Siglent SDM3055 Bench DMM  (Read 12596 times)
Offline alexvg
« on: August 10, 2019, 12:37:52 am »
During my SDM3055 reverse engineering... I've got 2 questions to you ?
Would you agree to pay a small amount for a new SDM3055 software (and maybe for the SDM3045) ?
http://www.pixule.com/39863606081_would-you-agree-to-pay-small-amount-for-new-sdm3055-software-and-maybe-for-the-sdm3045.html
Would you like to improve your SDM3055 by adding a some options ? (The SDM3045 may be compatible with these options.)
http://www.pixule.com/398636486944_would-you-like-to-improve-your-sdm3055-by-adding-some-options.html
Thanks for your help !
I'm working on a hack to show more digit to the Siglent SDM3055
I've got a source voltage of 2.5V (not a reference voltage)
Original : 240 000 points / 5 1/2 digits
[attachimg=1]
First hack result : 2 400 000 points / 6 1/2 digits
[attachimg=2]
The hack does not work correctly but give interresting result.
To be continued.
« Last Edit: August 19, 2019, 07:28:28 pm by alexvg »
The following users thanked this post: thm_w, tv84, kj7e, BillB, joeyjoejoe, umbro
Online tv84
  • Country: pt
Re: Hacking the Siglent SDM3055 Bench DMM
« Reply #1 on: August 10, 2019, 08:55:24 am »
 :clap:
Alexia, as requested. 2 of the latest app versions attached. Good luck!
« Last Edit: August 10, 2019, 04:16:04 pm by tv84 »
Offline alexvg
« Reply #2 on: August 11, 2019, 11:59:47 pm »
New results with 7.5 digits
The calibration data are not currently loaded :
[attachimg=1]
[attachimg=2]
[attachimg=3]
Offline Mr. Scram
  • Display aficionado
« Reply #3 on: August 12, 2019, 12:12:33 am »
Nice experiment! Not to rain on your parade, but are these numbers relevant? Do we know whether the 6.5 digit versions maybe have cherry picked or simply better references? I'm asking because many meters can show more digits but it's generally noise. Don't let it discourage you though and be sure to hack on!
Offline alexvg
« Reply #4 on: August 12, 2019, 12:23:57 am »
I've noticed some mistake into the code :
A real value of 1.499995 is displayed 1.4999 instead of 1.5000.
But a value of 1.4999995 is displayed 1.5000. It's because the rounding is done with a value of 5.0x10-7
« Last Edit: August 12, 2019, 12:25:58 am by alexvg »
Offline coromonadalix
« Reply #5 on: August 12, 2019, 12:46:22 am »
I would love to see this hack against a AC / DC / Ohms  calibrator ??
« Reply #6 on: August 12, 2019, 01:40:40 am »
Yes of course measurement stability can be questioned as SDM3065X uses the better LM399 reference.
Avid Rabid Hobbyist
« Reply #7 on: August 12, 2019, 07:26:30 am »
Could the 3065 and 3055  share the same design, minus the reference quality ??
Q: The hacked firmware, does it come from the 3065 ??
« Reply #8 on: August 12, 2019, 08:20:52 am »
Images from SDM3000 series thread
https://www.eevblog.com/forum/testgear/siglent-new-bench-dmm-sdm3055/
SDM3045X
SDM3055
SDM3065X
Avid Rabid Hobbyist
The following users thanked this post: coromonadalix
Offline alexvg
« Reply #9 on: August 12, 2019, 08:56:31 am »
The SDM3065X uses a LM399H (0.3ppm/C with 20ppm/1000h and 7uV noise) voltage reference and AD7175-2 "B" A/D (up to 24 peak-to-peak bits and 24 effective bits at 5 to 16 SPS).
The SDM3055 uses a MAX6325 "C" (0.5ppm/C with 30ppm/1000h and 1.3uV noise) voltage reference and AD7190 "B" A/D (up to 22.5 peak-to-peak bits and 24 effective bits at 4.7 SPS).
The SDM3045X uses TI REF5025K "High grade" (2.5ppm/C with 100ppm/1000h and 3uV noise ) and AD7190 "B" A/D (up to 22.5 peak-to-peak bits and 24 effective bits at 4.7 SPS).
Edit 2019-08-18
Difference vs SDM3055 :
- SDM3065X uses a separate power supply
- SDM3065X uses a LM399H voltage reference and a higher resolution ADC
- SDM3045X uses lower grade components but with a SDM3055 design (or SDM3055X-E PCB)
I suspect the SDM3055 could have a better ADC (see un populated U503 SSOP28) like ADS1256 (seems to be compatible)
I'm only using the original SDM3055 firmware.
« Last Edit: August 18, 2019, 10:08:44 pm by alexvg »
Online tv84
  • Country: pt
« Reply #10 on: August 12, 2019, 09:10:28 am »
If anyone with a 3045X wants to test the 3055 in his DMM, I can easily prepare an .ADS.
BUT, no responsibility for my side.
« Reply #11 on: August 12, 2019, 11:34:16 am »
To be continued.....................
And he busted it !  :palm:
PM pleading for help ...............
And returned.
Avid Rabid Hobbyist
Offline BillB
« Reply #12 on: August 12, 2019, 12:12:52 pm »
Well, that escalated quickly!   ???
Offline alexvg
« Reply #13 on: August 12, 2019, 01:36:40 pm »
I've done some bash correction to correctly use XML calibration file.
Now the value displayed is correct and the result is very interresting.
I'm investigate to display digits with a little lower font size.
With the input shorted to the ground :
[attachimg=1]
« Last Edit: August 12, 2019, 01:46:22 pm by alexvg »
Offline alexvg
« Reply #14 on: August 13, 2019, 08:59:19 am »
How to test the sdm-mod ? (only with the latest 1.01.01.19 firmware)
Step 1
Plug a formatted FAT32 USB Key (I use 4GB) into the SDM3055
Step 2
Connect to telnet (username:root, password:ding123)
Step 3
Backup all Siglent data into the USB key :
"cp -R /usr/bin/siglent /usr/bin/siglent/usr/mass_storage/U-disk0/"
After copying, unplug the USB key
Step 4
Plug the USB key into your computer
Copy "sdm-mod.app" into the "/siglent" of the USB key
Unplug the USB key
Step 5
Plug the USB key into the SDM3055 and go to telnet
Stop the current sdm.app application : use "ps" to list all application PID and kill the "./sdm.app" PID.
Generally it's the PID 550, you could kill it with command "kill 550"
Step 6
Prepare and run the mod
"mkdir /usr/bin/siglent/usr/mod"
"cd /usr/bin/siglent/usr/mod"
"cp -R /usr/bin/siglent/usr/mass_storage/U-disk0/siglent/ /usr/bin/siglent/usr/mod/"
"cd siglent"
"./sdm-mod.app"
Step 7
To stop sdm-mod.app, just press "Control C".
Restart the device to return to normal operation.
Step 8
To remove the mod folder
"rm -r /usr/bin/siglent/usr/mod"
sdm-mod.app 20190812 v01
SHA1 : 85ceab5b27570de364ab127d0c67a7deb17903ab
SHA256 : d26b360d6db9a89b058b4d44a2116c546b492e402df37fa1d491276a415fd5d0
Features
- Add 2 digits (7 1/2 digits)
Drawback
- On some menu the sign and/or the last digit is not display correctly
« Last Edit: August 13, 2019, 09:11:53 am by alexvg »
Offline alexvg
« Reply #15 on: August 13, 2019, 09:11:23 am »
I'll post later 2 other versions :
- sdm-mod.app with only one additionnal digit
- sdm-mod.app with two additionnal digits and some special text display (if it's work !!!)
I've not currently found the solution to change the font size.
It takes very long because my ARM debugger does not work...
If anyone know a solution to do a remote debugging of an ARMv7 GDBServer using Windows (or Debian)...
If you have a voltage reference, could you post your result ? (I use the BARmeter menu to display all digits)
« Last Edit: August 13, 2019, 09:14:13 am by alexvg »
Online tv84
  • Country: pt
« Reply #16 on: August 13, 2019, 09:54:58 am »
IDA?
Online Zorc
« Reply #17 on: August 13, 2019, 01:35:42 pm »
 :-+ Very cool you've got this far!
Interesting to find out how usefull this actually is!
I don't have a benchtop DMM but maybe in the future this might be a model I would consider buying (whenever I decide I really 'need' (read: want to have) one to supplement my handheld multimeters)
« Reply #18 on: August 13, 2019, 05:17:27 pm »
Has it been compared against a precision source  ??,  nice to have more digits,  are they worth displaying ??
Offline alexvg
« Reply #19 on: August 13, 2019, 05:45:06 pm »
I would like to know if additionnal digits are usefull or not.
The only thing I could say, there are an internal rounding error.
IDA?
I've tested IDA but it does not support this application and/or GDBServer integrated into the SDM3055.
Online tv84
  • Country: pt
« Reply #20 on: August 13, 2019, 05:46:35 pm »
What about via JTAG?
Offline alexvg
« Reply #21 on: August 14, 2019, 06:31:30 am »
Here's the 6 1/2 digits version :
SDM 3055 m1-1908-01
Feature
- Add 1 digit (total of 6 1/2 digits or 2 400 000 points)
SHA1
cb89541b22959c8cdae8a7a917b29ca7042d4bcf
SHA256
73bef60988a4b4ec684132795c0f6b99bb4fcf4d4e1824f0acad414d9f9f64cb
Step 1
Step 2
Step 3
Step 4
Unplug the USB key
Step 5
Step 6
"cd siglent"
"./sdm.app"
Step 7
To stop sdm.app, just press "Control C".
Step 8
Offline alexvg
« Reply #22 on: August 14, 2019, 10:13:37 am »
Never try to use SDM3045X or SDM3065X into the SDM3055
- The SDM3065X app program have an error loop
- The SDM3045X app program seems to work, but there no value on the display.
I'm currently testing a modification : I've removed the 0.5uV add/sub.
Edit: If you want to use the initial mod version I've posted (7 1/2 digits), You need to use the stat menu to display all digits and sign without any artefact.
I've noticed after a measurement of a battery (around 8V), using 20V range, there's a +/-20uV variation. This corresponding to a +/-2uV at the input of the ADC (with its 2.5V reference).
This seems to be very good.
« Last Edit: August 14, 2019, 10:17:06 pm by alexvg »
Offline alexvg
« Reply #23 on: August 16, 2019, 07:31:44 am »
I've add 2 little heat sink on voltage reference and ADC and I removed the fan.
The result is interresting (using 200mV range, short-circuit on the input and more than 4000 samples) :
- Previous noise result was 1.75uV
- Now the noise is reduced to 1.34uV (difference between min/max) and standard deviation is 0.207uV
I've mesured a -90dB noise into the 5V line... Why Siglent integrated the analog/logic power supply on the same board ? (except for lowering cost).
I'm thinking about adding a pre-rectifier/filter/regulator, changing the 5V regulator and probably changing the voltage reference to the LTC6655B-2.5 (if I could access to an excellent reference voltage).
I've received a AD584LH voltage reference, I've measured 20K samples at 2.5V with 20V range :
- 78uV span (+/-39uV)
- 11,1uV standard deviation
I think the SDM3055 seems to be more interresting than I though first (It only need some modifications to be more better).
« Last Edit: August 16, 2019, 05:15:41 pm by alexvg »
Offline alexvg
« Reply #24 on: August 16, 2019, 11:23:49 pm »
I've successfully running the remote debug using Texas Instrument Code Composer Studio...
Online tv84
  • Country: pt
« Reply #25 on: August 17, 2019, 12:32:49 pm »
Must study that.   :clap:
Offline alexvg
« Reply #26 on: August 18, 2019, 05:39:42 pm »
I've done a very long measurement to check the quantification level.
All measured values are using 22.5bits (slow mode or 4.7SPS). There's no loss during calculation.
22.5 bits corresponding to 6.77 digits. A full 24bits ADC (as seen in the SDM3065X) could provide 7.2 digits.
A long integration time with a sinc filter could provide more bits.
Offline alexvg
« Reply #27 on: August 19, 2019, 07:48:41 pm »
I need your help, and I've added 2 questions on the first post :
https://eevblog.com/forum/testgear/hacking-the-siglent-sdm3055/msg2605347/#msg2605347
I've done multiple meaures with an AD584 at 2.5V with 7.5 digits.
- Statistic averaging of 500 samples give 2.496551V
- Averaging of 23 values in 5 seconds give 2.49655078V
- Sinc filter of these 23 values give 2.4955098V
measurements of a 10K 0.1% 30ppm resistor with 7.5 digits
- Statistic averaging of 500 samples give 10.005007k
- Averaging of 23 values in 5 secondes give 10.0050084k
- Sinc filter of these 23 values give 10.0050086k
- Statistic averaging of 4k samples give 10.005006k
The noise level during A/D conversion is very low in Resistance mode (much more lower than DCV).
« Last Edit: August 19, 2019, 07:52:03 pm by alexvg »
Offline maginnovision
« Reply #28 on: August 19, 2019, 08:09:23 pm »
Are you comparing what you get to something that is actually more accurate, or are you just assuming this is giving you relevant data(not noise)?
Offline alexvg
« Reply #29 on: August 19, 2019, 10:27:56 pm »
Actually I trying to identify all most important noise sources (ADC, Voltage ref, op-amp, relay...).
I measure the distribution of data to see if internal processing maintain all available bits (from ADC).
The SDM3055 PCB analog-part is the base of the 4.5 and 6.5 digits DMM and I think it could be go to 7.5 digits (but with another ADC and some addtitionnal important modifications).
I don't know if 7.5 digits could be really possible with less minor modification, but I'm currently working on these little hardware modifications to check that (with software modifications and/or new software).
Offline alexvg
« Reply #30 on: August 20, 2019, 08:47:08 pm »
Sorry for double-post.
Online tv84
  • Country: pt
« Reply #31 on: August 20, 2019, 08:51:51 pm »
That's a wrong move!
« Reply #32 on: August 20, 2019, 09:05:15 pm »
You start charging and you might be asking Siglent to challenge you. Since you're modifying their firmware it could be a legal issue.
Offline joeyjoejoe
« Reply #33 on: August 20, 2019, 11:10:07 pm »
Asking for something to go wrong. Anyone needing more resolution would probably spend on the 3065 for supported 6.5 digits. Only hacker community would entertain this, and these are the folks who wouldn't pay for something like that.
Offline alexvg
« Reply #34 on: August 20, 2019, 11:20:01 pm »
That's not what I want.
My first tests was only done for evaluation only, I don't want to bother Siglent (I appreciate theirs products !).
I prefer providing my own firmware (not related to Siglent) but compatible with their product while offering additionnal, new or different features.
The SDM3055 is very good, I think it could be even better. My goal is to maximize this device.
You could be right... In the Android community some people are ready to pay some OS updates and advanced features for their old device. The SDM3055 is an old device.
I really appreciate your help, yours advices and opinions. Thank you.
« Last Edit: August 20, 2019, 11:26:29 pm by alexvg »
« Reply #35 on: August 20, 2019, 11:35:55 pm »
As far as I can see the 3055 is not old, it's a current product.
Offline blacksheeplogic
« Reply #36 on: August 21, 2019, 12:26:42 am »
OK, just so I have this straight. It's OK to steal but no OK to charge for stolen items? I was obviously under the misunderstanding that both were wrong.
If your providing a patch to 'extend' functionality (your own work, not enabling someone else work) to a device not sure where the legal issue comes from (sold or given). If your providing FW that's been patched which includes the original FW that's quite different as you are giving/selling someone else work.
« Reply #37 on: August 21, 2019, 12:43:01 am »
Both are wrong, but if you're charging people are more likely to take notice. It's not open source software, modifying it could be illegal.
To add some context Keysight Daniel bogdanoff once said they don't care if you hack your scope but they do care if you try and sell hacked scopes.
« Last Edit: August 21, 2019, 01:14:23 am by maginnovision »
Offline Mr. Scram
« Reply #38 on: August 21, 2019, 10:13:08 am »
He's saying he'd provide his own firmware not related to Siglent. Writing your own firmware and selling it is fully legal but not trivial to pull off.
« Reply #39 on: August 21, 2019, 10:39:04 am »
Is it me   or its going in the wrong direction ?   even if he provide his code,  where the code base came from ? Siglent   no/yes ?
Offline Mr. Scram
« Reply #40 on: August 21, 2019, 10:42:53 am »
Apparently not. Writing your own code completely from scratch is legal. Patching existing firmware with your own code also tends to be legal as you don't copy or provide someone else's code.
The following users thanked this post: thm_w, coromonadalix
Offline JohnG
« Reply #41 on: August 21, 2019, 01:52:37 pm »
This is really nice work, and the related discussion interesting as well.
But, there is a practical matter. A modified meter will have no warranty and reduced odds of support. Have you checked the prices of used 34401 meters lately? Not as many modern features as the 3055A, but they use a LM399 reference and do 6.5 digits without mods. The reference has a good chance of being aged and stable.
I have both meters. The 3055A is easier to use, for sure, but not much of a track record regarding accuracy. A modified version, even less so.
Just my $0.02.
John
"Those who learn the lessons of history are doomed to know when they are repeating the mistakes of the past." Putt's Law of History
Offline alexvg
« Reply #42 on: August 22, 2019, 09:23:16 pm »
The SDM3055 is based on Linux operating system with only a binary file written (partially) by Siglent launched during startup. Used internal libraries and all external libraries come from Texas Instrument, Linux system...
It's an old device from 2014... Lots of units have been sold : maybe some people would be interested in updating or improving their device without buying a new more powerful and perhaps more expensive DMM.
« Last Edit: August 22, 2019, 09:36:58 pm by alexvg »
Offline NoisyBoy
« Reply #43 on: August 23, 2019, 01:19:20 am »
Since you are trying to build something to sell, I doubt this is an appropriate discussion for this forum especially when you try to do a product interest survey.
The 3055 is not designed as an open hardware platform for open source firmware development.  I will never pay for “enhanced” firmware that does not come from the manufacturer, it is a sure way to void the warranty plus I will never trust the reading.
I am not saying it is a bad idea, just that it is more appropriate to be discussed in another venue.
Offline alexvg
« Reply #44 on: August 26, 2019, 07:10:36 am »
Thank you for this very interesting and informative discussion.
About software, I'm continuing working on the new firmware :
- I've successfully access to the display and keyboard.
- I'm waiting for the Lattice USB programmer.
- I'm currently investigate the protocol between digital and analog boards.
About hardware, I've received the components to build a pre-regulated power supply.
I'm waiting for the LTC6655BHLS8-2.5 voltage reference. I'll try one chip to measure ADC noise.
I'd like to build a 6-chip voltage reference : 6x LTC6655 in parallel to limit noise and deviation.
I'm waiting for a brand new calibrated voltage reference (based on the LM399AH).
I'm always searching for a calibrated LTZ1000ACH voltage reference module, I could not find it. Any ideas ?
« Last Edit: August 26, 2019, 11:09:22 am by alexvg »
Offline alexvg
« Reply #45 on: August 30, 2019, 04:22:37 pm »
I've found some interresting things :
- Only low-23 of 24 bits are used in the software (the 24th bit is used as a polarity bit - shame)
- I've done some precision simulations using datasheet ADC and I presume the ADC used have an internal precision of 24bits (instead of probably 32bits ADC used in the SDM3065X). Using software filtering and fast ADC sampling rate, it's seems to be possible to have a some better results.
- I've found some better Op Amp at the same price (ADA4528-2 instead of AD8629 and LTC1151 instead of AD706)
- I've received a LM399AH-based voltage reference calibrated at 10.00000V (my SDM3055X displays up to 10.001V)
- Resistance measurement (2 or 4 wires) is more stable than current or voltage measurement.
- The internal +5V, +15V and -15V have some 100Hz (from 50Hz AC).
« Last Edit: August 30, 2019, 06:54:56 pm by alexvg »
Online Kleinstein
« Reply #46 on: August 30, 2019, 08:04:01 pm »
It is kind of normal that 1 bit is used for the sign. So 23 bits + sign is kind of normal. Chances are the noise of the ADC is high enough, that there is very limited value to more resolution. With long time averaging a slightly high resolution can be possible, but this does not need the numbers from ADC chip itself.
The ADA4528-2 OP is not per se better than the AD8629 - they are different:  the ADA4528 is very low voltage noise, but higher bias and much higher current noise. The ADA4528 is good for signal sources with less than some 10 K impedance. However the AD8628 is much better for high impedance sources, like the input divider for the higher voltages.
In case one really want's to get sub 100 nV resolution, it would be more like an extra external preamplifier based on an OP like ADA4528. This could also add more gain and have a higher precision protection part (though possibly lower maximum voltage, more bias). Chances are in the lower cost meters the protection is just a resistor, maybe PTC. This could add enough resistance and noise so that it does not make sense to use the ADA4528.
Similar the LTC1151 and AD706 are good for different applications. So it depends on the use. AZ OPs can add quite some spikes, that might upset other circuit parts.
If resistance measurement is more stable than voltage and current measurements, this indicates that the reference noise is a mayor contribution. This might also be the external reference. A lower noise reference may be one of the few actually possible hardware mods, especially if the reference noise is really that bad. Usually the voltage readings near 0 V should be not effected by reference noise, but a reading near to the full scale (e.g. 2 V range, maybe 20 V range) should.
The chip SD ADCs are usually more like good for the lower end 6 digit meters, as the INL can be relatively high. The other problem is the limited range, so that there can be additional gain errors from resistors. So one may get more resolution (low enough noise), but the accuracy is still limited. These SD ADC chips are good for 5.5 digits, and they can even have a good point: the bandwidth for the input noise can be low, as they may not loose time to AZ phases.
The following users thanked this post: alexvg
Offline alexvg
« Reply #47 on: August 30, 2019, 08:25:57 pm »
Kleinstein> Thanks for your very interresting post !!!
I've intercepted data from the FPGA to get the used protocol.
Now I could convert the raw value into voltage value using the DMM calibration XML.
I've detected the real range of the analog input (VDC mode) :
- 200mV range : -312.33mV to +312.55mV (without ADC overflow)
- 2V range : -2.4984V to +2.5110V (without ADC overflow)
In VAC mode (100Hz sine)
- 2V range : 0 to +5.0356V (without ADC overflow)
Measurements done using Rigol DG4102 (as a DC voltage generator) and UNI-T181A (as a voltage measurement)
The VDC mode uses 23bits and 1bit for the polarity (sign)
The VAC mode uses 24bits but within a range from 2^23 to 2^24-1 (like 23bits)
The resistance mode (2W and 4W) uses the same limitation as VAC mode (23bits)
I'm also working on 2 mods in parallel : power-supply low-noise preregulation (plug and play) and multiple LTC6655BHLS8-2.5 chip for the voltage reference (need soldering).
« Last Edit: August 30, 2019, 09:10:30 pm by alexvg »
Offline alexvg
« Reply #48 on: August 31, 2019, 02:58:58 am »
It was hard to do but now I could access to ADC raw data.
Done :
- Read keyboard state
- Access to display
- Read ADC raw data
To do :
- Control capture mode
- Speaker
- External trigger
- VMC Output
Offline alexvg
« Reply #49 on: August 31, 2019, 08:20:08 pm »
I've done the investigation of the 20V VDC range... It's very funny :
-31.25V to +31.25V (without calibration)
Offline Nornand
« Reply #50 on: August 31, 2019, 08:52:04 pm »
Gainsetting seems to be 8 for the PGA.
Vref/8=0.3125V
Offline alexvg
« Reply #51 on: September 01, 2019, 02:04:09 am »
Yes the PGA is used... Shame.
The DMM is not an internal 240.000 points because it's possible to have up to 312500 points.
I've done a continuous decimation with a sinc function over 4096 samples captured with speed=FALSE. The result is similar to the speed=SLOW with 30sec statistics.
I confirm the fact that it's possible to have a super fast (with a low resolution) bar meter and a high precision (slow) display number simultaneously.
I'll try to enable a more faster tranfert between ADC, FPGA and ARM (4800SPS). I'll need to check if the current onboard Linux is compatible with critical realtime processing (or change to another Linux).
I've found the LT6657 voltage reference... Better than the LT6655...
Probably a stupid idea (or not) : If I'll build a LTZ1000AH voltage reference to supply multiple LT6657A-2.5. Does the final output voltage have a lower drift than a standard low-noise 5V voltage regulator powering multiple LT6657A-2.5 ?
« Last Edit: September 01, 2019, 02:12:39 am by alexvg »
Online Kleinstein
« Reply #52 on: September 02, 2019, 03:44:38 pm »
A standard 5 V regulator (if not getting too much heat and variable current) is good enough to supply a LT6655 or similar reference. The next step up would be a bootstrapped supply - though it should not be needed here.
In many aspects the LT6657 is not better than LT6655 (especially the LS8 version). Both are lower noise than the max6325, but the TC may be higher.
The big weakness of all these reference is not so much the noise, but the drift, hysteresis and sensitivity to humidity. So the difference between the MSOP and LS8 case may be significant.  Anyway for a 5.5 digit meter they should be good enough.
As the ADC needs the 2.5 V reference level, there are not that many better references to choose from. Several in parallel help with noise, but only little with long term drift rate.
The other limitation of the DMM design is that it is only +- 2.5 V (may be a little more, but not much) with high impedance.
So there are limitations to the design, that can not be circumvented.
Offline alexvg
« Reply #53 on: September 02, 2019, 08:07:59 pm »
The LTC6655 (LTC6655BHLS8-2.5) and LTC6657 (LT6657AHMS8-2.5) have a better long term drift than MAX6425 after more than 1000Hr and the noise is lower : the ADC used does not like the noise into its reference.
It's possible to use a LTZ1000A into the SDM3055 with a programmable voltage divisor (based on Vishay high precision low drift resistances). It's simple to do.
I study all possibility : After some little modifications on the device, I would like to know what level of precision it is possible to reach.
Offline alexvg
« Reply #54 on: September 04, 2019, 07:26:46 pm »
You're right ! Using the LT6657 will provide lower performance than the LT6655 (LT6655BHLS) with the AD7190.
So what to do ?
1. I could try to use up to 6x LT6655 ? (with a thermal isolation)
2. I could try to use up to 4x LM399A ? (but with a voltage conversion adding drift and low pass filter)
3. I could try to use one LTZ1000A ? (like LM399 with a similar voltage conversion adding drift and low pass filter)
Don't know what to do...
Online Kleinstein
« Reply #55 on: September 04, 2019, 08:40:41 pm »
The rest of the DMM (like gain stage and input divider - already used for the 20 V range) are still more suitable for 5.5 digits. So there is limited use in replacing just a a few points and hope to get much better performance.
It starts with having high impedance up to some 2.5 V only.
At some point is would be more like building your own DMM from scratch.
So I would consider anything more like a single LT6655 (with maybe some simple thermal regulation) overkill.
The main advantage of an LM399 would be better long term stability - still needs a very stable divider stage. It would also mainly help the 2 V range. The divider (+ buffer)  from a 7 V level down to a 2.5 V (maybe 3.5 V ?) level is probably more tricky than the LTC6655.
One also has to keep in mind that the power supply and thermal design may not allow much more power.
A change in the HW would mainly make sense if there is a simple, easy upgrade possible because of something like a design fault (could be a cost driven decision)  causing an isolated weak spot. Not sure if there is one such weak point -  chances are multiple source of similar size add up.
So a first point would be an in depth analysis of the performance of the original hardware. So how much of the noise is due to the input amplifier, possibly the input buffer, the ADC and the reference ?
For the drift this is more tricky as the result is valid for the one unit only.
Offline alexvg
« Reply #56 on: September 05, 2019, 06:19:27 am »
You're right, it's possible to use a 5V voltage reference but I need to check if this voltage is compatible with the analog input (AD7190 is compatible).
It's also possible to change the ADC... But I don't want to do this.
I'll need to check the noise from the input to the ADC (I'll need to build a LF amplifier probe).
I've also successfully set to 4800Hz the ADC and done a sinc filter to enhance the final resolution, the result is better than the original "Speed - Slow" (or 5 SPS)
« Last Edit: September 05, 2019, 07:18:25 am by alexvg »
Online Kleinstein
« Reply #57 on: September 05, 2019, 09:14:32 am »
I don't think a 5 V reference would be practica, as it's too far of the original 2.5 Vl. However a 3 V or 3.5 V (7 V / 2) ref. level could be working.
For checking the noise source, much can be done with the meter itself, reading data and look at the noise under different conditions. With s short, there should be no reference noise (at least not the normal one). With high gain in front chances are the amplifier noise would be dominant.
With the SD ADC there is however the ADC internal "gain"  (Usually actually only more frequent input sampling), that can make the analysis more complicated.
Is there an analog gain stage, or are they only using the ADC internal PGA stage for the 0.2 V range ?
Offline alexvg
« Reply #58 on: September 05, 2019, 07:13:03 pm »
It's possible to use a +5V voltage reference with the AD7190, but with the design of the SDM3055, I don't know...
Using VDC mode :
- 200mV - PGA = x8
- 2V - PGA = x1
- 20V - PGA = x8
- 200V - PGA = x1
- 1000V - PGA = x1
When I finish the logic analysis, I think I'll do an in-deep inspection of the analog part, and do lots of measurements.
At the moment I'm having trouble understanding the measurement of resistance of very large values (10M to 120M).
Online Kleinstein
« Reply #59 on: September 05, 2019, 09:32:10 pm »
For very large resistors a common method is to add some resistance like 10 M in parallel. This allow to measure very large resistors, though with reduced accuracy and a nonlinear function to convert from the raw reading to resistance.
As the SDM3055 is more like a budget instrument, I would expect this method.
Offline alexvg
« Reply #60 on: September 06, 2019, 09:17:31 am »
Yes, I've tested multiple methods without the correct results (I need to find the exact value displayed by the original DMM software). I continue my investigations.
Offline Nornand
« Reply #61 on: September 06, 2019, 10:46:39 am »
Datasheet says
Range:  100 MΩ              200 nA || 10 MΩ
Offline alexvg
« Reply #62 on: September 07, 2019, 03:01:28 am »
Datasheet says
Yes it's a current generator.
But I could not find the exact result.
Edit: I think I've found it... I need to do multiple checks.
« Last Edit: September 09, 2019, 08:35:16 am by alexvg »
Offline alexvg
« Reply #63 on: September 13, 2019, 02:30:46 pm »
I've finally found 2-wire and 4-wire measurement method.
I'm working to found how calibration work...
Offline alexvg
« Reply #64 on: September 14, 2019, 07:47:45 pm »
I'ts ok !!!
I've successfully found the method used for 2-wire and 4-wire resistance measurement.
I've discovered 2 problems with 2-wire and 4-wire 100M measurements :
- I've measured 144mV p-p 50Hz at the input of the ADC (with open input but with LO/Losense shorted)
- The XML calibration is not fully used : 0-50M segment is used for 0-120M and 50-120M is not used at all !!!
A interresting information :
- I've success fully generate a 0V ADC input with 0x800000 output data (with +/-1LSB precision during more than a minute- 7FFFFF to 800001)
I've done my measurements with :
- 10K to 130M with 0,01% 0,1% or 1% resistance value (for 2-wire and 4-wire)
- I use a Rigol DG4102 with filtered output and 20dB or 40dB attenuation for simulate resistance (only 4-wire)
« Last Edit: September 14, 2019, 07:59:04 pm by alexvg »
Online Kleinstein
« Reply #65 on: September 14, 2019, 08:15:00 pm »
It is not a surprise to find some hum for the high resistance values. Normally the high Ohms ranges should use relatively high voltage ranges. Here this seem to be the 2 V range. It may take some shielding to get good readings. Ideally the DMM would detect if there is too much AC background (e.g. to cause clipping in peaks).
144 mV_pp is not too bad, less than some 5% for the full range.  For a good measurement one may need to use a shield to keep hum away and maybe avoid using long unshielded cables.
For the 10 M range 4 wire Ohms measurement does not make much sense.
Offline alexvg
« Reply #66 on: September 14, 2019, 10:57:45 pm »
The voltage range could go up to 2V, it's current generator of (around) 200nA
I think you're right :
- Right because it's a nonsense to mesure high resistance value.
But I need to add something I'm thinking :
- Using a 4-wire cables it's possible to use shielded cables and differential input (for sense wires) for a more precise measurement.
But I could be wrong with this, I'm not sure (never tested, never simulated...)
I've also notice a strange math simplification inside the original software :
You could write :
f1(x) = 1/(1/x)
f2(x) = x
It's okay to say f1(x) = f2(x)
But it's false for a CPU !
Original Siglent Software uses a function like f1(x) that could cause a "divide by zero". But without simplifying the function, it's possible to avoid the "divide by zero" like f2(x). I don't understand why Siglent have done this...
Online Kleinstein
« Reply #67 on: September 15, 2019, 07:49:53 am »
The capacitive coupled mains hum during the high ohms measurement is present at the DUT and the hum is thus directly added to the voltage to measure.  So a differential measurement does not help here. The main things that helps is the usually good suppression of mains hum by the ADC, because it is integrating over a multiple of mains periods.
Shielding for the high ohms measurement can be tricky, as there may be more leakage. Some high end DMMs offer a special driven shield for ohms shielding, that reduced the voltage over the isolation and also help a little with settling.
I sometimes makes sense to measure high resistors, but in the MOhms range one does not really care about cable resistance and could thus use a 2 wire measurement.  Some meters only offer 4 wire Ohms for the lower ranges.
Using only 2 wire ohms reduces the cables to 2 and thus less leakage and less surface to pick up . So there is a point at high resistors when 2 wire ohms is more accurate than 4 wire ohms. It is not just the external leakage, but also leakage at the input protection.
Offline alexvg
« Reply #68 on: September 16, 2019, 09:28:37 am »
Measuring using 100M range with this device is a joke (using 4-w and 2-w).
The precision is awful (even with 50 and 60Hz filter)
« Last Edit: September 16, 2019, 10:51:28 am by alexvg »
Online Kleinstein
« Reply #69 on: September 16, 2019, 03:19:09 pm »
The accuracy gets worse with high resistors already due to leakage and bias currents (e.g. at the protection). With only a 2 V range (compared to often 5-10 V for other 6 digit DMMs) this also makes things more tricky. 4 Wire resistance is likely worse than 2 wire. The 100 M range is about 10 times less accurate than the 10 M range as there is the 10 M in parallel and thus only a relatively small change used.  Anyway getting 1% or so accuracy is not that bad with such high resistors.
A really accurate measurement would need a special instrument, more like an electrometer. This usually is with applying a fixed (e.g. 20 V or 100 V) voltage and than measure the current. This also need shielding and special, short cables. Also touching the resistors with the hand can change things from contamination.
Offline alexvg
« Reply #70 on: September 16, 2019, 07:02:15 pm »
My Applent AT2816 is more precise for high value... But it's a specific device.
I ve done 2 simple tests (with 2-wire and 4-wire)
- a 110M resistance measurement with 4-wire display 111,7M to 112,5M over a minute
- a 110M resistance measurement with 2-wire display 111,84M to 111,87M over a minute
A funny thing I see each time...
|SIGLENT SDG800 project        /dev/pts/0
SIGLENT Project http://         (none)
(none) login: root
Password:
Processing /etc/profile... Done
(SDG800 app and data are present inside the SDM3055 firmware)
« Last Edit: September 16, 2019, 07:04:19 pm by alexvg »
The following users thanked this post: exe
Offline alexvg
« Reply #71 on: September 18, 2019, 05:53:06 am »
I'm currently working on the frequency counter... It works up to 2,2MHz... Very limited ! My UNI-T181A (half the price) works up to 60MHz.
Now I could use the frequency counter.
I've measured a maximum frequency of :
- 1Vpp : 3,1MHz square and 2,9MHz sine (Range 200mV)
- 5Vpp : 4,1MHz square and 3,9MHz sine (Range 200mV)
- 10Vpp : 5,0MHz square and 3,7MHz sine (Range 2V)
- 20Vpp : 5,0MHz square and 5,0MHz sine (Range 2V)
« Last Edit: September 21, 2019, 09:13:38 pm by alexvg »
Offline alexvg
« Reply #72 on: September 22, 2019, 06:33:09 pm »
Now I've got the Capacitance measurement process.
It's possible to display a value with 5, 6, 7 or 8 digits... But the accuracy is very low...
To check capacitors, I use a LCR Applent AT2816 at 100Hz / 2V
I've done 34 measures with a list of 18 capacitors from 100pF to 10000uF.
All capacitors have been checked before and after the measure of the SDM3055 with the LCR.
Minimal error : 0,04%
Maximal error : 10,2%
Average error : 2,2%
Std Deviation : 0,0228%
The capacitance measurement of the SDM3055 is not very useful than a real LCR.
I've also checked all capacitors using a low-cost LCR (Peaktech 2170). The results are always better than the SDM3055 !
Offline alexvg
« Reply #73 on: September 23, 2019, 06:23:20 am »
I've selected some components to build a pre-regulated power supply.
- A couple of LM2991/LM2941 is used to create the +/-18V (regulated later for +/-15V)
- A single BA80BC0WT is used to to create the +8V (regulated later for +5V)
- A 100V/2A/0.7V Schottky bridge rectifier (+/-15V line)
- A 40V/2A/0.5V Schottky bridge rectifier (+5V live)
The main goal is to decrease the noise of these power lines and to reduce the heat of the main regulators.
The selected LDO regulators have a switch pin which will be connected to the internal switch signal.
I also need an additionnal power supply because I'll use a LTZ1000ACH instead of the MAX6325 for the ADC voltage reference.
Offline CDaniel
« Reply #74 on: September 23, 2019, 06:54:50 am »
If the big error is for high value electrolytic caps it is fairly normal for a multimeter ... even two different Fluke don't display the same value and could be easily 10% off . For the simple method  of charging/discharging the cap the time is dependent on losses and current used ...
« Last Edit: September 23, 2019, 07:03:05 am by CDaniel »
Offline alexvg
« Reply #75 on: September 23, 2019, 07:56:59 am »
The error is spread over all the measures (but above the Siglent specs).
The current used is specified into the SDM3055 datasheet (200nA to 1mA).
I don't understand why Siglent doesn't incorporate a little frequency generator to do a more precise and FAST capacitor (and inductance) measurement using the 2 or 4 wires
« Last Edit: September 23, 2019, 10:24:39 am by alexvg »
Offline alexvg
« Reply #76 on: September 24, 2019, 06:20:07 am »
I've done DCI and ACI. It was very funny because DCI 200uA range have an internal inverted polarity (-11uA measure and +11uA display).
Offline alexvg
« Reply #77 on: September 24, 2019, 08:42:17 pm »
Here is a list of reverse engineering features done :
- DCV (all range, filter mode and 10M/10G for 200mV and 2V)
- DCI (all range and filter)
- ACV (all range)
- ACI (all range)
- Resistance 2-wire (all range including 100M !!!)
- Resistance 4-wire (all range including 100M !!!)
- Frequency (all ACV range)
- Capacitance (all range)
Remaining functions :
- Diode
- Continuity
- Temperature
Edit: All functions done !
« Last Edit: September 24, 2019, 11:45:22 pm by alexvg »
The following users thanked this post: coromonadalix, umbro
Online tv84
  • Country: pt
« Reply #78 on: September 25, 2019, 09:38:46 am »
 :clap: Where there's a will there's a way!
Online Kleinstein
« Reply #79 on: September 25, 2019, 01:33:24 pm »
An inverted polarity could happen if they use a trans-impedance amplifier for the low range. In this case I would expect to have at least 2 such ranges, possibly the option to add a 20 µA range.
Offline alexvg
« Reply #80 on: September 26, 2019, 06:18:13 am »
It's possible to have more range, but I need to do a noise measurement (FFT analysis of the ADC running at 5KHz).
It's also possible to easily replace the used ADC with a better one (like the AD7175-2 used in SDM3065X with some additionnal electronic).
The SDM3055 board is more interresting than the SDM3045 and SDM3065, modifications are easy but available space inside is very limited.
Offline alexvg
« Reply #81 on: October 02, 2019, 09:42:58 pm »
This is just a first test for the new interface
[attachimg=1]
The following users thanked this post: ebclr, tv84, coromonadalix, Qw3rtzuiop, JohnG, BillB, umbro
« Reply #82 on: October 03, 2019, 02:27:52 am »
Puke green ?  :scared:  :-DD
Otherwise nice improvements.  :-+
Avid Rabid Hobbyist
Offline e0ne199
  • Country: id
« Reply #83 on: October 03, 2019, 01:30:15 pm »
dang you should use your ability to hack 121GW instead of this DMM...
Offline alexvg
« Reply #84 on: October 25, 2019, 09:23:46 pm »
I'm currently writing the software using a Windows emulation to have an easier development.
I'm thinking about 2 hardware mods :
1. pre-regulated power supply (I already have components but haven't got time to assemble them)
2. remove ADC and ref and add an additionnal board with a better reference (LTZ1000A) and a better converter (AD7175-2 or AD7177-2)
But i don't know if the pcb and current components provide a noise floor enough lower to enable a good enhancement.
I haven't got the tools to measure the 0.1-10Hz noise at very low level.
« Last Edit: October 25, 2019, 09:26:23 pm by alexvg »
Online Kleinstein
« Reply #85 on: October 25, 2019, 10:24:41 pm »
Measuring the noise if the DMM is not that difficult. The main tool it the DMM itself.  The first test is measuring the noise with a short at the input and than calculate the noise from the data. The second test needs a low noise external reference to measure a voltage at a significant part of the full scale. This gives the second noise case that includes the noise of the DMM internal reference and maybe some extra ADC parts.
The 0.1 to 10 Hz noise would be approximated by getting data with 20 SPS and calculate the RMS value over some 10 seconds. Chances are one could use a slightly different sampling rate (e.g. 25 SPS) and thus a slightly different frequency range throughout the tests.
For the reference a 7 V reference needs an extra stage to divide down the reference to the ADC level (e.g. 2.5 V maybe 3.5 V). This can add quite some uncertainty / drift. So it is questionable if it is worth using an highly stable LTZ1000 with a divider that adds quite some drift. The 20 V range would still use the front end divider that adds quite some drift and also some noise. For a 2.5 V reference there may be a simpler replacement with another 2.5 V ref chip.
For the ADC, lower noise ADCs may need a lower impedance input signal. As far as I understood it, they use the internal gain for the low ranges (200 mV, current). So the new ADC also needs this option.  So the change is not that simple.  One may have to build a small separate PCB for the Reference and ADC - so first step towards building your own DMM.
The input amplifier (buffer) may be limiting the noise performance. I would expect some AZ OP as a buffer -  a compromise between noise and input bias. At least an AZ buffer should be sufficiently linear.
Offline alexvg
« Reply #86 on: November 19, 2019, 09:29:57 am »
Some news :
I'm working on an interface using a HTML/CSS rendering for a more scalable and flexible possibilities/features.
It takes a very long time because the power of the main CPU is very limited: I need to do a high level of optimizations and remove some CSS features (who need too much process time/power).
I don't know when the first alpha/beta version will be release, "probably" in 2-3 months.
Hope you'll enjoy my work.
The following users thanked this post: tv84, CDN_Torsten
Offline supperman
« Reply #87 on: November 26, 2019, 06:46:05 pm »
Quick Questions.. As I'm considering a Siglent DMM:
Why don't they integrate current flow over time to get Amp/H. Seems super easy.. to get them to measure total charge on a battery for instance. (all the data is there). Seems that I would need to log all the readings and then run it in Excel on a PC? Does the Siglent log both Voltage AND current over time? (or only one or the other using the PC software).
@alexvg any thoughts on expanding the math to do integration math? Also, given the time you are spending on this.. why not just design a DMM? Siglent will never let you sell software for their device..
Offline BillB
« Reply #88 on: November 26, 2019, 07:54:42 pm »
Good question regarding integrating current.  I don't believe the EasyDMM software supports dual measurement mode at all.  For meters without a channel card option, it only performs a single measurement function.
Offline alexvg
« Reply #89 on: January 06, 2020, 01:35:04 pm »
Hi, sorry for the delay...
This Siglent DMM only do a single a measurement at a time.
Current/Voltage uses time slicing with a relay.
I'm working on a solution to provide multiple length and simultaneous math integrations.
The following users thanked this post: purpose
Offline klausES
  • restore Sony ES/Esprit era
« Reply #90 on: February 05, 2020, 12:25:31 pm »
Hi there,
I think I read somewhere that the temperature (at which the factory was calibrated) was 23.5 degrees Celsius?!?
(I'm not sure anymore, can't find the text anymore).
If I remember correctly, it would be the temperature at which the voltage reference and the relevant part of the board later
most closely correspond to the factory calibration in measurement mode.
Are there measurements or empirical values by how much these values in practice, e.g. 30 or 35 degrees C device (inside) temperature could differ
(to what extent the temperature compensation corrects this or not)?
The question has a background (a thought), but I'd rather wait and see if someone answers.
regards klaus. "Art is when you can't do it ... because if you can, it's not art"
Online Kleinstein
« Reply #91 on: February 05, 2020, 02:14:55 pm »
How much the readings can change with temperature is sometimes noted in the instructions. Usually some +-5 K is included in the base accuracy  for more I would expect something on the order of 2-15 ppm/K as an upper limit. The lower ranges (200 mV and 2 V) that don't use the input divider can be better, as the reference used and the ADC have a relatively low TC. The input divider for the higher ranges (20 V, 200 V,...) is expected to have lower accuracy, due to the TC of the divider that is more in the 5-15 ppm/K range.
I would not expect much of an extra correction in software, as this would need an extra step at the calibration.
Offline klausES
« Reply #92 on: February 05, 2020, 03:14:17 pm »
Thank you for your answer.
Had the following considerations.
It is clear that if constant values ​​are to be compared, it would be advantageous to keep the instrument at the same temperature at all times.
If it is almost all about constant (much more than the best absolute values),
it would not be advantageous to maintain a constant but higher temperature than the constantly fluctuating ambient temperature.
A precisely controlled temperature that is always x degrees higher than the ambient temperature
is much easier to implement inside the device than a possible cooling to keep it the same.
In winter my room has e.g. 18 degrees, but also 30 degrees in midsummer (or sometimes just> 30 degrees, unfortunately no climate).
If the device were now "always from the start" kept at, for example, exactly 30 degrees
(controlled by a precisely temperature-controlled heating and in a control loop proportional fan speeds),
fluctuating room temperatures in the range of 18 to approx. ... 29 degrees would hardly have any changing effects on the temperature in the device.
hence the question:
How much would the absolute values ​​at a constant 30 degrees differ from those of the factory calibration at 23.5?
Edit:
corrected a text error that led to confusion (see above).  :-\
« Last Edit: February 05, 2020, 07:09:00 pm by klausES »
Online Kleinstein
« Reply #93 on: February 05, 2020, 04:22:12 pm »
The TC specs for the ADC and reference are at some 1 ppm/K. The actual performance may be a little better than this at moderate / common temperature. So the 6.5 K higher temperature my change the reading by some 5-10 ppm, so hardly detectable with a 5 digit meter.
The ranges with input divider and current ranges are likely less accurate so more like 10 times the change.
For the 5 digit resolution there is not much need for a constant temperature. This would be something for a 7 digit meter. At that level a oven for the reference (e.g. LM399 or LTZ1000) is the normal solution.
Offline imo
« Reply #94 on: February 05, 2020, 05:18:41 pm »
hence the question:
In case you have a temperature sensor mounted in the box, you may simply compensate the readings based on that internal temperature. I did it with my meter at 10V range (34401A, I had to mount the sensor).
The absolute accuracy of the temperature sensor is not important here. A resolution of 0.1C would be enough for your meter, imho.
Then you will calibrate against that internal temperature. The DMM starts cold, and you will be logging the external ref voltage readings against the internal temperature - as the internal temperature of the DMM rises it will sweep up the temperature for you. You will get a "Vref" voltage vs. internal temperature dependency.
And the rest is math like y=ax+b (or whatever your dependency is like) within the internal temperature range of interest.
The internal temperature of your meter (after it stabilizes, it may take hours) is always Tdmm = Tx + Tambient, where Tx is a constant (an assumption here is your meter is not internally thermostated).
« Last Edit: February 05, 2020, 06:00:15 pm by imo »
Offline klausES
« Reply #95 on: February 05, 2020, 07:19:19 pm »
Thanks for the explanation.
I will take a closer look at the behavior and test the suggestions when the device arrives (hope tomorrow).
At first I only came up with the idea because the fan will get a temperature-controlled control right from the start (possibly also a quieter fan type).
Then a regulated "very specific selectable and then constant" higher temperature than the fluctuating environment would have offered itself.
Online Kleinstein
« Reply #96 on: February 05, 2020, 07:42:38 pm »
A temperature regulated fan is a two sided thing. With changing air speed temperature gradients can change - this can have more effect than the overall temperature change.  It really depends on the circuit if a temperature controlled fan helps of makes things worse.
Offline klausES
« Reply #97 on: February 05, 2020, 09:22:14 pm »
Thanks for the information. I can understand your objections regarding fluctuations.
I'll pay attention to that.
In any case, a linear series regulator is planned here (neither pwm on the power side,
nor something clocked in the measurement and control).
There will also be no "fluctuations" by simply switching on / off,
Instead, an adjustable minimum speed (always active as a base load), an adjustable incline, a delta t> 1 kelvin
and a limitation of the maximum speed (moderate, far below the possible full speed of the fan).
Offline killingtime
« Reply #98 on: February 21, 2020, 11:28:31 pm »
Really like the new interface. It's an improvement on the standard.
Putting the measurement units after the measured value (e.g. mV) in larger typface like Keithly have done makes it easier to read.
Making the measured value digits larger would be an idea - if there is space on the display. Means you can read the meter from further away.
Well done.
Offline alexvg
« Reply #99 on: February 29, 2020, 05:40:50 pm »
I'm currently working hard on the HTML/CSS rendering engine.
I don't know when I could release a test version.
It's a long way...
Edit: You could try these mods :
6.5 digits : https://www.eevblog.com/forum/testgear/hacking-the-siglent-sdm3055/msg2614659/#msg2614659
7.5 digits : https://www.eevblog.com/forum/testgear/hacking-the-siglent-sdm3055/msg2612406/#msg2612406
« Last Edit: February 29, 2020, 05:45:42 pm by alexvg »
The following users thanked this post: purpose, coromonadalix, killingtime, CDN_Torsten, umbro, brucklerniklas
Offline stefkpl
« Reply #100 on: April 18, 2020, 05:45:42 pm »
I will try this week. It's seem to be a good work :)
Tkx
Offline klausES
« Reply #101 on: April 18, 2020, 06:58:14 pm »
What options are there to explicitly save the existing calibration file
(the still unchanged original factory calibration) of an SDM3055 ?
Offline alexvg
« Reply #102 on: April 18, 2020, 07:03:42 pm »
Since the start of the health crisis in France and around the world, I have put this project on standby.
I currently have to manage 7/7 server clusters (from several websites) so that children can work during confinement.
I will resume the development of this project at the end of confinement or when I have free time again.
Take care of yourself and your loved ones.
Alexia from France.
The following users thanked this post: purpose, tv84, Qw3rtzuiop, JohnG, not1xor1, wolfy007
Offline Electro Fan
« Reply #103 on: April 23, 2020, 05:01:36 am »
Nice Work alexvg!  Hope all goes well and that we see you back on this project.
tautech, maybe you could encourage Siglent to hire alexvg or retain alexvg's services.  Sure looks like a lot of creative and capable product management and product development horsepower that could benefit Siglent, their products, and their customers.
Offline pipe2null
« Reply #104 on: April 24, 2020, 07:57:17 pm »
"Me Too":  Yes, great work alexvg!  I hope things calm down in the world soon so we can all get back to kewl projects like this one.
I have an SDM3065X-SC and want to try out UI and 7.5 digit FW mods if/when it's possible to do so.  I considered attempting the 7.5 digit hack per the instructions, but since I have a different model, I thought I'd ask first to avoid messing things up.  Is it reasonably "safe" to try out the 7.5 digit mod on a 3065?
Offline rwgast_lowlevellogicdesin
    • LowLevel-LogicDesign
« Reply #105 on: May 03, 2020, 03:04:04 am »
I have an sdm3045x and am pretty satisfied with it for the most part but, the resistance measurement 2 and 4 wire only measure to 3 digits beyond the decimal in its lowest 600 range, how hard would it be to hack a fourth digit? I'd be willing to upgrade my voltage reference to the max one used in the 3055 if need be!
Offline Muessigb
    • My blog
« Reply #106 on: May 19, 2020, 11:30:59 pm »
Some news :
Why put a full HTML/CSS rendering engine on an embedded device. That's really inefficient. I don't like when people do it on Desktop, but it's just awful on embedded.
You can always use SDL or QT embedded to do the GUI.
I've done my own fair share of aftermarket firmware and SDL usually works quite well.
Offline Elasia
« Reply #107 on: May 20, 2020, 12:37:12 am »
It's a different main board, i wouldnt normally say this but you got the top of the line model, i'd leave it be since its the flagship less you are out of warranty or the like
You can cross application files across same mainboard revisions safely, beyond that you are on your own, its the reason you can turn the new SSA 3000X Plus into a 3.2Ghz SVA model, same main board, just a different application file
Offline pipe2null
« Reply #108 on: May 20, 2020, 01:31:11 am »
True on the hardware.  Beyond that it depends on how they architected the firmware/software if the UI code is well separated from the underneath bits.  Most of the firmware hacks I've seen posted are reasonably safe and reversible to try but only for a very specific model.  After I posted I finally saw a note from OP of the "thou shalt not" try on other models...   :palm:  Was hoping it might be possible to safely kill normal UI process and try out OP's much better UI.  Odds of that actually working are low, but thought I'd ask.  I'm not up to speed enough to hack around with it myself yet, need to get through a few more learning curves with JTAG and the like.
(*Sarcastic*) And thanks a lot for bringing up the SSA3000X+ "upgrade".  It took months to convince myself to wait until next year before digging into the RF side of things and getting related "field upgradable" SA/VNA toys...  -ER-, I mean essential tools.   >:D
...  SVA?  Sounds like I'm behind on reading.  But I really, really shouldn't catch up.  Bad idea.  Bad, bad idea.  I don't go to TEA meetings.  I don't have a problem.
Offline alexvg
« Reply #109 on: July 15, 2020, 04:20:47 pm »
Hello,
Just a little message to thank you for your support.
I have been sick with Covid-19 since June. Despite all the precautions taken (mask, distance, hand washing ...), that was not enough.
It's very difficult everyday with a lot of symptoms. The healthcare team is great but clueless.
Please protect each other, take care of yourself.
Offline Pinkus
« Reply #110 on: July 15, 2020, 04:57:13 pm »
 :wtf: We all wish you a speedy recovery. Take care of yourself and get well!
Online tv84
  • Country: pt
« Reply #111 on: July 15, 2020, 05:06:32 pm »
Hello,
I wish you well!
Offline ZhuraYuk
  • Country: ua
« Reply #112 on: August 24, 2020, 07:53:46 pm »
Hello,
I wish you a strong health and good luck.
I want to thank you for this awesome work. 3055 is very capable device. Running 6-1/2 digit mod, I checked 4 wire resistance against 10OOhm 0,002% resistance standard. Only last digit was noisy same as my freshly calibrated Keithley 2100.  Also I compensated wire resistance because it was a little 0,0136 Ohm off when 4 wires shorted. After connecting to reference resistor the reading was stable up to last digit. All measurements were dine after 2 hours warm up.
Offline uski
« Reply #113 on: September 30, 2020, 08:11:00 am »
Anyone has news of Alexia ? I just found this thread and the last message is months old, so I am worried now !