| Rien:
Hi I have a Hantec 2c42 but now after update it is changed in an 2D82 I like to go back to my original 2C42, but how? When i try to update again the meter is seen as an 2D82 an FPGA V255 |
| Microcheap:
Try using the options "Force update ARM" or "Force update FPGA" or simply manually install the correct FW using DfuSe: https://www.eediscuss.com/forum.php?mod=viewthread&tid=13676&fromuid=26055 By the way, did you managed to fix the multimeter? |
| Rien:
Thanks I try that. The meter is working as before yes , see my answer above. I have take a look at it, first try was no succes, but after that I have downloaded both files and then force update with local files. That was the solution. Everything is now ok also the resistance measurement. |
| Pinkus:
Caution, my Kaspersky shows a Trojan-warning when I tried to enter the eediscuss-site. Though it might be a false alarm, but I would not bet on it. |
| Microcheap:
--- Quote from: Pinkus on January 29, 2020, 07:42:04 am ---Caution, my Kaspersky shows a Trojan-warning when I tried to enter the eediscuss-site. Though it might be a false alarm, but I would not bet on it. --- End quote --- Probably a false alarm, I never had a problem there. I also checked it with Virustotal.com and it came clean: |