Author Topic: Hantek 6022BE Maximum voltage in 1x and 10x mode  (Read 5265 times)
Offline kavehmTopic starter
  • Country: ir
« on: February 26, 2024, 01:30:20 am »
What is the maximum voltage for 1x and 10x mode?
35V voltage is written next to the BNC connector. Is this maximum voltage for 10x mode?
« Last Edit: February 26, 2024, 08:04:13 am by kavehm »
Online Aldo22
Re: Hantek 6022BE Maximum voltage in 1x and 10x mode
« Reply #1 on: February 26, 2024, 09:02:16 am »
The following users thanked this post: kavehm
Offline horo
    • My GitHub Projects
« Reply #2 on: February 26, 2024, 10:53:17 am »
Aldo22,
thank you for the link to my older post, so I don't have to repeat myself :)
Maybe Dave's video is also interesting
Martin
  • Country: ir
« Reply #3 on: February 26, 2024, 11:45:08 am »
You can find information here:
https://www.eevblog.com/forum/testgear/hantek-6022be-20mhz-usb-dso/1100/
Be careful!
As I understand, in 1x mode, the safe voltage is 5 volts and it can be used up to 35 volts.  And in 10x mode, the safe voltage is 35 volts, which can be measured up to 350 volts.  Have I understood correctly?
Offline horo
« Reply #4 on: February 26, 2024, 01:55:13 pm »
The usable input voltage range is -5V ... 5V. Voltages below -5V are limited to -5V and voltages above 5V are limited to 5V, so it makes no sense of applying higher voltages, this is just a safety margin.
Voltages outside the safe range of +-35V (can) destroy your device.
With a 1:10 probe the limits are 10 fold, i.e. measure between -50V..+50V, the safe range is theoretically -350V..350V. However, I would suggest keeping the limit of +-50V as values outside this range will not give useful results. Also check the safe operating voltage of the probes. Please be very careful with high voltages.
You can check the principle schematics of the oscilloscope in the user manual on page 20:
https://github.com/OpenHantek/OpenHantek6022/raw/main/docs/OpenHantek6022_User_Manual.pdf