Author Topic: bug found? Hantek DSO5102P: different measures in AC vs DC coupling  (Read 3423 times)

0 Members and 1 Guest are viewing this topic.

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 7817
  • Country: us
Re: bug found? Hantek DSO5102P: different measures in AC vs DC coupling
« Reply #25 on: April 16, 2021, 10:35:28 pm »
And this is consistent across all channels?

A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline radiolistener

  • Super Contributor
  • ***
  • Posts: 3342
  • Country: ua
Re: bug found? Hantek DSO5102P: different measures in AC vs DC coupling
« Reply #26 on: April 17, 2021, 04:11:27 pm »
And this is consistent across all channels?

yes, the behavior is exactly the same for both channels. Both have the same frequency response.
 

Offline DL2XY

  • Regular Contributor
  • *
  • Posts: 75
  • Country: de
Re: bug found? Hantek DSO5102P: different measures in AC vs DC coupling
« Reply #27 on: April 18, 2021, 10:08:56 am »
Your mains voltage is not a sine wave but a trapezoid.

Simulating the measurement with a clean trapezia clearly shoes the effect:
During the constant voltage phases the coupling capacitator is charging, in the transients this charge is transported to the other half-cycle resulting in overshoot.


 
« Last Edit: April 18, 2021, 10:15:40 am by DL2XY »
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 16604
  • Country: us
  • DavidH
Re: bug found? Hantek DSO5102P: different measures in AC vs DC coupling
« Reply #28 on: April 18, 2021, 10:12:47 pm »
There are plenty of oscilloscopes which do not suffer from additional distortion when measuring distorted AC line waveforms.  With a single pole response, even the difference between 1x and 10x probe attenuation yielding 10 and 1 Hz cutoffs is not visible.
 