@TheoB -- thanks for replying with thoughts and insights! Much appreciated! Let me update with some further details.
Assuming the boards output is 50 ohm, I would just use SMA to SMA cable and set the DSO's input to 50 ohms.
Unfortunately the output of the DUT is not designed for a 50ohm load.
And apologies: I should have been explicit about the input termination.
- The input is 1Mohm // 13pF (Keysight CX1151A Passive Probe Interface Adapter)
You mention loading is an issue, so you need a X10 probe. That still has 15pF as load...
Yes, I want the DUT to see a high impedance so I thought to use a 10:1 probe instead of 1:1 probe, but I did not consider the capacitive load that is there.
To add more detail: I'm not selecting a 10:1 probe for the voltage attenuation; the voltage being measured will be <10V. A 1:1 probe would be better for less noise, but such probes have considerably less bandwidth.
...Above 20MHz you can better use a resistive divider into 50 Ohm coax cable. That keeps a constant 500 Ohm load up to high frequencies. But you have to manually set the attenuator factor in the scope.
This sounds interesting as a solution! 500MHz bandwidth would be sufficient.
The attenuation can be configured.
Given the input termination (1Mohm) I could make a 2:1 cable from coax with a 1M resistor in series with the "probe tip" (SMA connector).
How to address probe compenation?
I have not seen an SMA style resistive probe before. Do they make such a product?
Neither have I; it does not seem to be a thing.
Ultimately I'd like a solution that's more robust than the slide on adapters; avoid the probe tip being damaged and less clumsy than multiple adapters.
It sounds like a custom cable from coax might be the way to go.