Ugh... close, but no cigar...
Look where the current shunt is, no way that's good enough clearance for 8kV.
I think the cutouts and spaceings are on the other end of the fuse, where the input is.
So you only get separation if the current is high enough to blow the fuse.
Great design!
/sarcasm
LOL. Step back......
What is the voltage over shunt and fuse when fuse is not blown?
8kV? Or 100-300mV ?
8 kV limit applies to voltage mode only.
Current mode has spec of how much kV it can withstand for 1 min
when fuse is blown, and that is usually 1 kV (or max voltage in voltage mode)...
Critical spec for current mode is max breaking current and fact whether it will, in fact, extinguish the plasma and break current as opposed to having a breakdown and having arc discharge accident.
Specifications are quite complicated and there are many caveats when reading what is what..