Author Topic: Help! I bricked a perfectly functioning 2467B!  (Read 3293 times)
Offline vintageradiobuffTopic starter
« on: April 04, 2024, 07:56:30 pm »
Hello all,
I have posted this on the Tek group on groups.io, but I am posting it here too for a wider audience.
I had a perfectly functioning 2467B scope (1989 vintage). After reading about possible loss of calibration data due to the NVRAM battery dying, I decided to put in a new battery (mine is the older version, below 50000 serial #) which has the battery separate from the NVRAM. The battery had a date code of 2689, so the battery is 35 years old. I did hook up a backup battery before removing the Lithium battery and confirmed that pin 28 of the RAM chip had at least 3 V on it from the backup battery after removal of the original battery. After putting in the new replacement battery, I also changed out the 4 electrolytic capacitors on the A5 board. When I put the scope back together and fired her up, I got the dreaded " Test 04 Fail 13" message indicating corrupted or lost calibration data. Not sure how this happened since I hooked up a backup battery before removing the original battery. So, the very thing I was hoping to prevent happened by my being proactive!
Before doing all this, I had made a copy of the RAM contents by running through the Exercise 02 option, so I have the original calibration data available. My question is: how to write this data back to the RAM?  (my scope does not have the GPIB option).
I look forward to your suggestions and hope to get this fine scope functioning again!
Offline alpher
Re: Help! I bricked a perfectly functioning 2467B!
« Reply #1 on: April 05, 2024, 11:08:55 pm »
Can't help you with writing to ram, (went through the same ordeal as you a few years back) . But nonetheless my advice, don't be affraid of "calibrating" the scope back.
Since you didn't change any parts nor moved any trimpots all you have to do is to go through is a couple of CAL procedures, even doing it first time in my life it took me ~an hour and a half.
Not hard at all. CAL-1 and CAL-2 if i remember it right, and of these only the first one takes any time.
Online colorburst
« Reply #2 on: April 06, 2024, 12:03:27 am »
I like Mark's suggestion to write cal data using customized firmware. Sounds like he already wrote the program, so all you'll have to do is burn a set of EPROMs and temporarily install them in the scope. Hopefully they're not soldered down!
Offline DavidAlfa
« Reply #3 on: April 06, 2024, 12:42:20 am »
No sense, you need a programmer to burn the eeproms, then just program the nvram in it!
Same happened to me. It was working, but after replacing so many exhausted NRAMs I decided to make a preventive maintenance, however it died after desoldering and placing it into the programmer.
But you'll need a programmer. Though a SRAM is very easy to interface, any Arduino, AVR, PIC, STM32 board with enough pins will be ok.
Try writing the data you got from EXER02 to 0x1FE00 - 0x1FFFF.
I recall doing this, but not 100% sure, I ended making a complete recalibration, everything was way off after 30 years.
https://www.eevblog.com/forum/index.php?topic=5397.msg4747154#msg4747154
« Last Edit: April 06, 2024, 12:58:33 am by DavidAlfa »
Online colorburst
« Reply #4 on: April 06, 2024, 02:05:22 am »
OP mentioned their model has an external battery, so the SRAM needs to be programmed while installed in the scope, unlike a Dallas block.
Offline DavidAlfa
« Reply #5 on: April 06, 2024, 05:20:16 am »
Ah, sorry, missed that detail.
Offline MarkL
« Reply #6 on: April 22, 2024, 07:51:35 pm »
So, the mentioned firmware utility to restore the EXER 02 data is done.  If you're not a watcher of the big 24xx thread, it's been posted here:
  https://www.eevblog.com/forum/testgear/tektronix-2465b-oscilloscope-teardown/msg5463367/#msg5463367
Online colorburst
« Reply #7 on: April 22, 2024, 08:34:23 pm »
Thank you Mark and Don, great work! This closes the loop on cal data woes for 2400 A/B scopes.